Answer / srinivasu & phani
Calibration is only performance check for instrument, where
as System Suitability is the criteria in which we look for
various parameters like analyst, reagents, system for a
particular analysis.
| Is This Answer Correct ? | 27 Yes | 1 No |
Answer / rahul kamble m.pharm
CALIBRATION is just to check whether the instrument is work
according to SOP procedure or not? it is done for
minimising the errors in the reading.
whereas SYSTEM SUITABILITY is process in we check various
parameters like analyst, its qualification, reagent for
whether they are AR or RS grade. its standard, calibration
of glasswares, instruments. LAB. CONDITIONS ETC.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / bhushan
Calibration is only performance check for instrument, where
as System Suitability is the criteria in which we look for
various parameters like analyst, reagents, system for a
particular analysis
| Is This Answer Correct ? | 6 Yes | 1 No |
Calibration is only performance check for instrument, where
as System Suitability is the criteria in which we look for
various parameters like analyst, reagents, system for a
particular analysis.
along with that calibration is done on perodicali butyou
must run sst in every analysis
| Is This Answer Correct ? | 3 Yes | 1 No |
Answer / manoj k. harne
In calibration we check the performance of instrument only by using primary or reference standards. But, in case of system suitability we checks over all performance of the analytical method against chemical reagents used, analyst,glassware,environment conditions. But, both activity closely related with each other. If instrument is not calibrated well then system suit can not meet criteria.in short calibration is precision of instrument and system suit is precision of proposed/intended method/procedure for analysis.
| Is This Answer Correct ? | 2 Yes | 0 No |
calibration means whether the instrument properly working
or not, when compared to reference standard based on some
parameters. if it fails it is a quality impacting insident.
system suitability means it is use full for analyst,
reagents, columns. these are intended use.
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / ch s raju
calibration is the performance check of the instrument
where as system suitability is your system is suitable for
that method we follow some ich guidelines for system
suitability parameters like resolution,RRT etc
| Is This Answer Correct ? | 2 Yes | 3 No |
Answer / narasaiah
both are the same if calibration is ok system suitability
parameters pass.when the entire system configeration is in
good condition then only system suitability will pass other
wise not pass
| Is This Answer Correct ? | 2 Yes | 10 No |
what is internal standard define
2 Answers Apex, Coromandel International Limited, Hikma Pharmaceuticals, IPCA,
what is mean by UOD?
Generally for Normal Phase HPLC,stationary phase is Polar.In place of polar if we take non-polar means what happened & why?
principle of UV , PDA , RI detector in HPLC
how can we determine RRF value in HPLC FOR CALCULATION OF IMPURITY ?
Ratio of Water and Acetonitrile(diluent) is not recommended to use in cleaning method validation,why?
0 Answers Aurobindo, Reddy Labs,
How they found 1mL of K.F reagent is equivalent to 5mg of water and if we change the composition of K.F reagent, is it can neutralize more amount of water?
What is the Deference between Bulk density & Tap density ?
2 Answers Hikma Pharmaceuticals, Neuland, Reddy Labs,
why we conclude the area% for the chromatographic purity?
1 Answers Amneal Pharmaceuticals, Lupin,
While performing TOC sst analysis Zero shift disabled & sample analysis zero shift enabled why?
How to decide the standard concentration for related substances hplc method
Describe 5 components of SFC ?
1 Answers College School Exams Tests,