What should be the calibration procedure for wavelength
accuracy in case of HPLC?
Answers were Sorted based on User's Feedback
Answer / ankama rao.puppala
wavelength accuracy was done by the following reagents
1.Caffeine
2.Erbium Perchlorate
3.benzophenone
In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...
In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.
| Is This Answer Correct ? | 17 Yes | 0 No |
Answer / s.elumalai
Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / aslam
Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / alpesh patel
Can you give any reference from pharmacopoeia? with chapter number...
| Is This Answer Correct ? | 0 Yes | 0 No |
Why we check moisture,ash & AIA in product
Why we are using the methanol in Kf titration?
What is the difference between qualitative and quantitative analysis (hplc)
How to establish pKa for a new API molecule ?
2 Answers Neuland, Sun Pharma,
defination of wavelength & wave number
if partition coefficient value was positive means, what it indicates ?
How will u perform degradation study in detectors other thann PDA in HPLC Methodology
how can we calculate void volume?give its importance in brief
How to choose Normal phase or Reverse Phase HPLC method Conditions for given sample?
In the HPLC Calibration done wavelength accuracy done between 200nm-280nm .but not done remaining 300-400nm not done ?
how to develop the icp ms method? Application of icp ms?
why we are using in ph calibrations ph4,7,9.2 buuffers only?
3 Answers Reddy Labs, Unichem,