What should be the calibration procedure for wavelength
accuracy in case of HPLC?

Answers were Sorted based on User's Feedback



What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / ankama rao.puppala

wavelength accuracy was done by the following reagents

1.Caffeine
2.Erbium Perchlorate
3.benzophenone

In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...


In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.

Is This Answer Correct ?    17 Yes 0 No

What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / s.elumalai

Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.

Is This Answer Correct ?    2 Yes 1 No

What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / aslam

Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity

Is This Answer Correct ?    2 Yes 1 No

What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / alpesh patel

Can you give any reference from pharmacopoeia? with chapter number...

Is This Answer Correct ?    0 Yes 0 No

Post New Answer

More Analytical Chemistry Interview Questions

plz tall me that any pharmacopeia and guideline is par meted to use 1.68,4.01,7.00,9.00,12.45 buffer for ph meter calibration

2 Answers   Torrent Pharma,


difference between HPLC & LC-MS

9 Answers   Lambda,


sop of a uv visible spectrophotometer double beam elico model

0 Answers   Sun Pharma,


what is hold time in Gc?

1 Answers  


How to determine the EDTA content by potentiometry titration in Ceftriaxine sodium

0 Answers  






what is the purge flow & how to calculate

0 Answers   Granules,


What is the difference between ROI & sulphates ash?

1 Answers  


what buffer you can select for acidic nature analyte and how ?

2 Answers   Dr Reddys,


in karl fischer reagent which solvent is used for dissolving iodine sulphur dioxide and imidazole

4 Answers   Strides,


what is the criteria for choosing capillary column in GC??is it dependent on sample load?

3 Answers  


why did uyou choose analytical chemistry?

1 Answers  


wat is reverse phase in hplc

2 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)