Answer / ankama rao.puppala
wavelength accuracy was done by the following reagents
1.Caffeine
2.Erbium Perchlorate
3.benzophenone
In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...
In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.
Is This Answer Correct ? | 17 Yes | 0 No |
Answer / s.elumalai
Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.
Is This Answer Correct ? | 2 Yes | 1 No |
Answer / aslam
Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity
Is This Answer Correct ? | 2 Yes | 1 No |
Answer / alpesh patel
Can you give any reference from pharmacopoeia? with chapter number...
Is This Answer Correct ? | 0 Yes | 0 No |
plz tall me that any pharmacopeia and guideline is par meted to use 1.68,4.01,7.00,9.00,12.45 buffer for ph meter calibration
difference between HPLC & LC-MS
sop of a uv visible spectrophotometer double beam elico model
what is hold time in Gc?
How to determine the EDTA content by potentiometry titration in Ceftriaxine sodium
what is the purge flow & how to calculate
What is the difference between ROI & sulphates ash?
what buffer you can select for acidic nature analyte and how ?
in karl fischer reagent which solvent is used for dissolving iodine sulphur dioxide and imidazole
what is the criteria for choosing capillary column in GC??is it dependent on sample load?
why did uyou choose analytical chemistry?
wat is reverse phase in hplc