Answer / gaurav oza
step-1 Organoleptic test- color, odour, shape.
step-2 test of functional groups.
step-3 confermatory test for functional groups.
step-4 qualitative tests for presence of compounds.
step-5 IR spectroscopy
step-6 UV spectroscopy and titration methods for
quantitative tests.
step-7 HPLC for better results and impurity study.
| Is This Answer Correct ? | 19 Yes | 1 No |
Answer / dhananjay s chavan
Pre method development requirements
1 Literature search
2 Published method
3 Availability of Pharmacopiel/Supplier/DMF Impurities
4 Availability of Pharmacopiel/Supplier/DMF Method
5 Workability of Pharmacopiel/Supplier/DMF Method
6 Identify different source of API
7 API degradation data from Supplier/DMF
8 API stability data from Supplier/DMF
9 Calculate imp level as per ICH for FPS
Method development
10 API Solubility in different pH
11 API Solubility in different solvent
12 Placebo interference
13 Main peak recovery/peak shape/response
14 Degradation study in different condition with
autoclave
Achieve degradation between 5 to 20%
Peak purity of all known impuriite(s) and main peak
Peak purity of all unknown impurities above LOQ
level
Check the elution of all degradants by extending
runtime
15 Establish LOD/LOQ
16 Filter study
17 Elution of all source impurities in final method
18 Reproducibility of method in different system and
different column lot
19 Check System suitability requirement (Resolution,
Similarity factor, Std RSD…etc)
Parmeteres to be done on finlized method
20 RRF Development in different system
21 LOQ, 100% and 150% Recovery in duplicate
22 Linearity
23 Stability sample and different batch analysis
24 Solution stability
25 Robustness in final method (Flow/temp./pH variation)
26 Method equivalency between final method and other
method(s)
27 Method evaluation by validation team
| Is This Answer Correct ? | 3 Yes | 9 No |
What is the meaning of Base deactivated vials
What is the limit for single maximum impurity limit as per USP??
Why we use Uracil for HPLC calibration in Detector wave ength accuracy test?
Why Deuterium lamp is used as UV light Source for UV.
why we use actone for Gpv test in hplc calibration
Why water not directed by FID detector
2 Answers Amoli Organics, Lupin,
Why do we need GC instead of HPLC? or Why do we need HPLC instead of GC?
Which Zone of our country like conducting of Stability studies(Accelerated & Long term stability studies)as per ICH guidelines.
in HPLC injection volume is increase ,R.T will increase ?
5 Answers Claris Lifesciences,
why are using toluene in uv calibraion in resolution performance
when we talk about change in entropy we consider total entropy means of surround +entropy of system but when we talk about change in Gibss free energy we consider entropy of system.explain with reason.
What is the purpose & importance of using the polystyrene film in ir calibration?