In the test for control of absorbance in UV calibration why
do we use only potassium dichromate and what is purpose of
taking a specified amount of 57.0-63.0mg?

Answers were Sorted based on User's Feedback



In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / manikatna

potassium dichromate is highly colourproducing agent
this amount is as per EP And edqm monograph.
if you change concentration change specific abs values

Is This Answer Correct ?    78 Yes 17 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / bhagya lakshmi.k

It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm

Is This Answer Correct ?    53 Yes 4 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / sunil.m.kulkarni ( avon/arch p

Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concen­tration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectro­photometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs

Is This Answer Correct ?    36 Yes 5 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / sagadevan

becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,

Is This Answer Correct ?    18 Yes 4 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / devidas

it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity

Is This Answer Correct ?    18 Yes 7 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / vivek

we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...

Is This Answer Correct ?    12 Yes 5 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / srinathsagar

UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)

Is This Answer Correct ?    4 Yes 1 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / pvinod6447

because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm

Is This Answer Correct ?    6 Yes 4 No

Post New Answer

More Analytical Chemistry Interview Questions

why we get negative values in uv when i placed sample which has some concetration but not in standard

1 Answers   Nirma,


What is the meaning of 624 in DB-624 Capillary column used in GC?

5 Answers   Aurobindo,


Why linearity of detector performed in HPLC calibration

2 Answers   Cipla, Intas,


how to convert normal phase hplc to reverse phase and vice versa?

1 Answers  


How does GC did work?

1 Answers  






For limit test of heavy metals in BP, Method C require that the substance is ignited at a temperature not exceeding 800 °C. Why confines such the temperature?

0 Answers  


why we use caffeine for hplc calibration?

5 Answers   Aurobindo,


inhouse product is in capsule form in combination and RLD is in tablet form then can we proceed for multimedia CDP? in inhouse capsule product disso is paddle with sinker in release media is there then RLD product in tablet form then with same as paddle with sinker we can proceed n.a.?

0 Answers  


what is the difference between qualitative analysis and quantitative analysis and explain them?

1 Answers   Dr Reddys,


why in hplc peaks are in positive side(it means upside) not in negative side?

3 Answers  


how to regulate noise in hplc ?

2 Answers  


why we use Dichlorobenzene.nitrobezine.t-butyldi sulhate.for calibration of gc Head space

0 Answers   Raks Pharma,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)