Answer / bhagya lakshmi.k
It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm
| Is This Answer Correct ? | 53 Yes | 4 No |
Answer / sunil.m.kulkarni ( avon/arch p
Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concentration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectrophotometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs
| Is This Answer Correct ? | 36 Yes | 5 No |
Answer / sagadevan
becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,
| Is This Answer Correct ? | 18 Yes | 4 No |
Answer / devidas
it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity
| Is This Answer Correct ? | 18 Yes | 7 No |
Answer / vivek
we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...
| Is This Answer Correct ? | 12 Yes | 5 No |
Answer / srinathsagar
UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)
| Is This Answer Correct ? | 4 Yes | 1 No |
because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm
| Is This Answer Correct ? | 6 Yes | 4 No |
why using Potentometry What Purpose? differance between manual and potentiometry? why using Some of materials in usig potentiometry? What is use of mericuric acetate in Potentiometry
which one is better separation in high pressure liquide chromatogram orlowpressure chromato gram in hplc
why 0.005 M sulphuric acid is used to calbrate spectrophotometer and why caffiene is used for calibrating hplc
2 Answers GVK, Torrent Pharma,
Why a1% value is used for some product ? What is the criteria for selection a1% value ?
why phepolphthalene giving pink color when we titrate against to NaOH Solution ,giving colrless titrating aginsiting with acids?
2 Answers Graviti Pharmaceuticals, Vimta Labs,
How we can identify process related and degradation impurity in single method with short period?
What is the exact reason to use THF and IPA in buffers of Reverse phase HPLC as a solvent?
How to know HPLC column performance?
what is classification of elemental impurities? what is risk assement in elemental impurities?
why reference flow required during hplc analysis on rid?
in determination of(acetone) moister content we are using methanol replaced by pyredine.what hapend in that reaction.
If compound Pka is 5 then which buffer you will select ?