Answer / alpesh patel
Capillary column : Stationary phase is coated with the
inner wall of the column & it is
having high lenghth and low internal
diameter.
Packed column : Stationary phase is directly coated in
the column & it is having short length
and high internal diameter.
| Is This Answer Correct ? | 116 Yes | 17 No |
Answer / mohite vikas
Capillary column gives better performance as compaired to
pack colums
| Is This Answer Correct ? | 61 Yes | 19 No |
Answer / balla reddy. p
Packed columns are more durable, has higher sample capacity
(Sample load) ,more phase selectivity, and has lower cost.
The disadvantages are larger amounts of bleed, smaller
number of theoretical plates/efficiency (i.e. separating
power for similar compounds is much low) and particle loss
into the detector. So it can not be coupled with MS or other
sensitive detectors.
Capillary columns have the advantage of larger theoretical
plate numbers /efficiency /higher resolution and inertness.
They excel at low ppm to ppt level analysis. They are
limited in phase selectivity options but their low carrier
gas use to the detector and low bleed characteristics allows
Mass Spectroscopy and other sensitive detectors to be used
effectively.
The use of bonded phases permits higher oven temperatures to
be used in applications.So capillary column withstand higher
column oven temperature compared of packed column.
| Is This Answer Correct ? | 45 Yes | 4 No |
Answer / pharmainterview
Packed columns: Have high particale size silica (old)
Capilliary columns: Have low particle size silica (new)
Regards
pharmainterviewquestions.blogspot.in
| Is This Answer Correct ? | 18 Yes | 3 No |
Answer / m m shaikh
Capilary columns:-1) Capillary columns gives high resolution ,Accuracy in analysis ,Rapid results than tha packed .
2)
This is bcoz of High lengh & low innr diametr.
It has 10- 100 meter lengh & 0.2 to 0.53 inner diametr.
3)This columns gives highr theoratical platea,inertness.to samples.
4)packed columns have low length & high id.
| Is This Answer Correct ? | 12 Yes | 0 No |
Answer / swapnil borse
Packed columns have higher sample capacity than capillary
columns, although the difference has been greatly reduced by
the large-bore 530µm capillaries invented by Agilent.
Improvements in detector sensitivity have also reduced the
need for large samples. The one area in which packed columns
may have an advantage is in analysis of gas samples.
For almost all other samples, capillaries provide much
better efficiency (narrow peaks) which leads to greatly
improved peak separation. In fact, the separating power is
so great that many analyses can be done on surprisingly
short columns in very brief runs. This time saving
translates directly into reduced turnaround time and
increased sample throughput.
For new or updated methods, we recommend capillary columns
unless there is some overwhelming reason for using packed
columns.
| Is This Answer Correct ? | 14 Yes | 9 No |
Answer / ksreddy
capillary column does nothaving any stationary phase. its
contain length and diameter.it is used for only back
pressure in system calibration.
packed column :its contain stationary phase.
| Is This Answer Correct ? | 13 Yes | 65 No |
which effect is mainly responsible for chromatographic separation when silica is the statinary phase at TLC? thanks in advance
give clarity of linearity and range in method validation
a bottle of concerntrated HCL (36.5) was labelled as 37% (w/w) and density 1.19m/l.What is hcl molarity?
What is zero water in TOC Analysis?
gc Calibration % RSD failed what is the possible reasons?
we calibrate HPLC once in 6 months while iam doing calibration i came to know that calibration was failed. then how can i justify that the samples analyzed on the system during this 6 months was wright? how can i forward in to the next step for those batches?
Why PH range from 0to14.
2 Answers College School Exams Tests, Reddy Labs, Wings Customer Care,
what is the Bracketing & Matrixing.
defination of wavelength & wave number
Which buffer is used for LC/MS mobile phase & at What Conc.
what is the film thickness in ftir
How you can calculate pKa ?