why the same column give different retention time in hplc?
Answers were Sorted based on User's Feedback
Answer / nabil
there are many resons for different RT
1-The pressure used(because that affects the flow rate of
the solvent)
2- The nature of the stationary phase(not only what
material is made of, but also particle size.
3-the exact compositionsof the solvent.
4-the temp of the column.
| Is This Answer Correct ? | 23 Yes | 0 No |
Answer / prabhakar
Because preparation of mobile phase variation(mostly)
| Is This Answer Correct ? | 6 Yes | 4 No |
Answer / dhaval bhatt
1)column temp
2)presser not stable
3)leakage column from any side
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / vaibhavgoswami
due to pressure ,
different column packing,(silica)
different maker
| Is This Answer Correct ? | 0 Yes | 0 No |
if mobile not sonicated properly the retention time will vary
| Is This Answer Correct ? | 1 Yes | 3 No |
Answer / phanindra
I think it is due to contamination of HPLC column remedy is column kept for wash with 100%polar solvent.
| Is This Answer Correct ? | 0 Yes | 6 No |
Why Urasil and Caffeine use for HPLC wavelength accuracy calibration. which is most suitable, Why? other than these which can be used?
What is the major difference b/w Drift and Noise
4 Answers Aurobindo, Dr Reddys, Reddy Labs,
WHY given much noise PDA dector then VWD Dector
How to fix the sample concentration in related substances development.
3 Answers Novartis, Watson Pharmaceuticals,
why water, methanol, ethanol ets solvents are used in UV SPECTRIPHOTOMETER for blank reading ?
How does GC did work?
Why 162 mg NaOH used for 1N volumetric solution in 150 ml water then 54.5 ml diluted with 1000 ml
in each media we require to use SLS? how to proceed?
what is %labelled amount in content uniformity of dosage unit and its calculation?
what r the reasons to get peak telling?
how do you perform reverse enginearing study for drug containing sodium chloride sodium citrate sodium bicarbonate
which are the sizes of capsules?