One pH means how much volts and milivolts?
Answers were Sorted based on User's Feedback
Answer / s b singh
59.12 mV is equivalent to 1pH. this may check in USP also
| Is This Answer Correct ? | 58 Yes | 4 No |
Answer / pratap rasam
Ideally as per Nerst Equation 1 pH unit=59.5 mv. But we are
all the time dealing with non-ideal situations only and
thus following table should be dealt with
pH mv
1 357.0
2 297.0
3 238.0
4 178.5
5 119.0
6 59.5
7 0.00
8 -59.5
9 -119.0
10 -178.5
11 -238.0
12 -297.0
13 -357.0
14 -416.5
| Is This Answer Correct ? | 28 Yes | 5 No |
Answer / arash.babakhanian
pH MEASUREMENT IS HAPPENED USING THE DIFFRENCE BETWEEN TWO
POTETIAL VALUE CAUSED BY THE POTENTIAL RELATED TO PROTON
EXCHANGE BETWEEN INTERNAL SOLUTION AND GLass hydrated
membrane and the potential between the external solution
and glass hydrated membrane that is called bindry potential
| Is This Answer Correct ? | 10 Yes | 12 No |
Why are use silicon oil mr?
stability study is going on up to 3years for famotidine-USP but the pharmacopeia is revised and tlc test is replaced with hplc test. At the 4th year frequency the product is failed in hplc test .how can i assain the expiry or retest date and can i stop or continue the programme?
In UV calibration we use Holmium perchlorate solution made in 1.4M perchloric acid The question is that 1.4 M perchloric acid we make is it in water or glacial acetic acid?
How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain
0 Answers Graviti Pharmaceuticals, Vimta Labs,
how does h2so4 prevent oxidation of mohr salt before it is titrated?
Why water not directed by FID detector
2 Answers Amoli Organics, Lupin,
What is Arsenic Test?
0 Answers Flamingo Pharmaceuticals,
Why we select scan range from higer wavelength to lower wavelength in uv visible spectroscopy and ftir spectroscopy ? .
Why we more noise in PDA detector compared to UV detector?
Why the same type of columns showing different RT for a single product?
What are the primary standards?
could negative ions be produced by bombardment process in mass spectrometry?