Answer / ramalingeswararao
APPLYING THE WATER FACTOR AT 25 DEG CENTIGRADE 0.99608.
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / jitender axiva
Calibration procedure is nothing but comparision of one of
the know magnitude value erf. as std to the test one.
in the case of glassware std calibarted glassware must be
10 times (least count)higher accuracy than test unit.
e.g. if we want to calibrate 10 ml pipette then we must
have to calibrate it through pipeete which of 1 ml with
least count less than that of test one.
| Is This Answer Correct ? | 2 Yes | 3 No |
Answer / mr.bujji reddy kanchi
Designated volume, mL 10 25 50 100 250 500 1000
Limit of error, mL 0.02 0.03 0.05 0.08 0.12 0.20 0.30
Limit of error, % 0.20 0.12 0.10 0.08 0.05 0.04 0.03
Weight measurement by taking water against standard
glassware(class-A type)
| Is This Answer Correct ? | 2 Yes | 5 No |
what is the value of pH
how do start hplc method development? how much sample required for developing the method development?
3 Answers BMS, Cipla, Indoco Remedies, Sun Pharma,
What is a difference between potency and purity?
in stress study if your api not soluble in hcl naoh h2o2 then what require to do?
Is it necessary to validate both assay and related substances for usp official methods for products.
Loss of linearity in tkn analysis, possible reasons?
why you are using benzoic acid for water content determination in TEA
In Related Substances Test how we can reside the limit of impurity ?
why the sucrose used in calibration of polarimeter
signifcance of detector lenearity in calibration f hplc
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
In the HPLC Calibration done wavelength accuracy done between 200nm-280nm .but not done remaining 300-400nm not done ?