Is it required to inject 6 injections for % RSD of a system suitability in lc or gc
Answers were Sorted based on User's Feedback
Answer / naveen
In LC:
5 replicates of standard if %rsd is 2.0% orless
6 replicates of standard if %rsd is more than 2.0%
Is This Answer Correct ? | 8 Yes | 3 No |
Answer / borse
Yes, it is necessary to inject 6 injection it show your system repeatability (precision), if it is within limit, it means your system is ready to proceed.
If any correction please suggest.
Is This Answer Correct ? | 4 Yes | 0 No |
Answer / rk
As per USP we can inject 3 to 6 , but %RSD vary as per number of injections
Is This Answer Correct ? | 4 Yes | 0 No |