Is it required to inject 6 injections for % RSD of a system suitability in lc or gc
Answers were Sorted based on User's Feedback
Answer / naveen
In LC:
5 replicates of standard if %rsd is 2.0% orless
6 replicates of standard if %rsd is more than 2.0%
| Is This Answer Correct ? | 8 Yes | 3 No |
Answer / borse
Yes, it is necessary to inject 6 injection it show your system repeatability (precision), if it is within limit, it means your system is ready to proceed.
If any correction please suggest.
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / rk
As per USP we can inject 3 to 6 , but %RSD vary as per number of injections
| Is This Answer Correct ? | 4 Yes | 0 No |
What is difference between known and unknown impurities
Home heating oil or furnace oil what is made from? Can (base oil plus diesel/kerosen) = home heating oil?
difference between AAS & flame photo meter ?
How can we calculate offset value pH measurment
Why potassium dichromate was using in Uv calibration?
why ph electrode is storing in KCL solution?Why the pH max is 14.0?
Why we are using only KBr in IR spectrometer why not KCl
In BP2013, Loperamid HCl monograph. Assay by titration with 0.1N sodium hydroxide using hydrocloric acid 0.01N and reading the volume added between the 2 points of inflexion. I have a question that if the diluent solvent is ethanol is certainly consumed a amount of volume of titrant, so this volume must be eliminated on the result calculation or not apart from first point which is subtracted above.
what is the dipolement and how it is related to IR
1 Answers AJ Square Inc, Aurobindo,
What is meaning of audit trail in hplc
3 Answers DRL, Lee Pharma, Raks Pharma,
what are the standard methods for HPLC?
What is the definition of RRF