Answer / anil
There is no Integration function in the conventional UV-Vis spectrophotometer which
accumulates the signal. For example, the total required time will be 1000 sec. for 1000 data
points and it takes 1 sec. to measure one datum. In this case, all 1000 data have the same signal
to noise ratio (S/N).
But in a PDA instrument which has a 1000 photodiode array, 1000 data points can be
measured in 1 sec. and it would take 1/1000 sec. to achieve the same result obtainable in 1 sec.
in a conventional instrument. Therefore, when the same sample is measured for 1000 sec in a
PDA instrument, the signal is accumulated and is 1000 times greater than when measuring for 1
sec. The noise will be 10001/2
. This means that the S/N ratio is improved by 10001/2
.
This resulting benefit of fast data acquisition is termed Felgett’s S/N Advantage or
Multichannel Advantage
| Is This Answer Correct ? | 10 Yes | 1 No |
in which situation require to take incident in validation?
How many clauses in ISO!
what is assay in hplc
When recording an IR spectrum in the lab, we take no special precautions to account for the presence of oxygen, nitrogen, or argon in the air. What are the consequences of this?
Is it nessesary all multimedia dissolution require descriminatory?
what is the dipolement and how it is related to IR
1 Answers AJ Square Inc, Aurobindo,
In the HPLC Calibration done wavelength accuracy done between 200nm-280nm .but not done remaining 300-400nm not done ?
I got 88.0 % assay for a drug in HPLC analysis for same injection i got 40.0% related substances is it possible? what is the relation between results of assay and purity?
what does pH stand for?
How can we Qualify a unknown impurity by HPLC
What is difference between Potency & purity
is a bonded water molecule in an organic material can be easlisy and accuractly determined by KF rather than LOD