Answer / carl_ellis
Let's assume base metal elements only, since noble metal
elements are so expensive that they are under lock and key;
adequate protection from loss of identification.
For types T, J and K, (and by exclusion, E) the options are
-a spectropscopic assay which will positively reveal the
alloys' composition, but given that the cost of an assay is
proabably 20x times the value of a base metal element, it's
not likely to happen.
- visual identification of a type T by the copper color of
its positive leg which is copper. The alloy leg is a shiny
'metallic' color.
- testing Types J or K with a magnet.
The Type J positive leg is made of iron which has a very
strong magnetic attraction.
The Type K negative leg is made of a nickel alloy which is
mildly magnetic.
Type E is non magnetic, but so are other types.
It can be tricky to distinguish mild attraction on thin
gauge wire.
Testing with a multimeter can be very misleading, because of
the difficulty in determining both the hot junction AND the
cold junction temperatures. Without accurate temperature
values, of what use is a mV reading?
Carl Ellis
Measure First
| Is This Answer Correct ? | 5 Yes | 0 No |
Why 4ma - 20ma signal used in instrumentation??
what is the difference between profibus and controlnet networking.
How to calibrate a orifice differential pressure flow transmitter in process?
how to measure level under ground tank ? which type of transmitter very accuracy ?
what is difference between positioner and position transmitter in control valve
I have installed a RTD whose length is 2250mm(RTD stem length), but it's upper 200 -250mm is out of the reactor, and this upper 200mm stem is surrounded by atmospheric temp. And the lower portion i.e. 2000mm is in contact with the measuring fluid, then does the final measuring temp. Gonna change as the upper 200mm RTD stem is out of reactor?
1 Answers Process Control Technologies,
how to do dynolab wind speed sensor loop test, how to simulate value,i have simulate mA 4-20mA but not responding in DCS faceplate, please kindly advice me any one.
if the t/c o/p is 4-20mv then how plc recognize it because we are dealing with the 4-20ma in the plc ????????????????
what is air to open and air to close control valve and difference btwn this...and how to measure milli amp by this...and what is the role of positioner and i/p converter in this..plz help me
use of split range ? where it is used ? in which control loop it is used ?
how can be calibrate flow indication transmitter (rosemount) it shows excess milli amps nearly about 24 ma on full scale, but its scale was set on 20ma on full scale..???how can be simiulate..???
why k factor is important to know in flow transmitter. if we don,t know k factor what will be issue we can face .
1 Answers Flour Tech Engineers (P) Ltd,
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)