different between calibration range and instrument range ?
Answers were Sorted based on User's Feedback
Calibration range at which the
instrument calibrated
instrument range:maximum range of
instrument at which we can calibrate
instrument for example if we have a
pressure tx instrument range 0~10bar
we can calibrate it at any position
between 0~10bar suppose we calibrate
it at 0~4bar its mean its calibration range
is
0~4bar and instrument range 0~10bar
| Is This Answer Correct ? | 26 Yes | 1 No |
Answer / raja
CAlibration range is the range upto which we have marked
readings on a guage ( for e.g) .
Instrument range is the range upto which instrument can
detect measurand .
calibration range is always less than Instrument range.
Dead zone and hysteresis may take position between the
difference in instrument and calibration ranges
| Is This Answer Correct ? | 3 Yes | 0 No |
What is difference between scada and virtual instrumentation?
1)What is the suitable MOC for flow meters working at temperature between 200 - 250 cen. 2)In which application the mass flow meters are not advisable 3)Is it advisable to use radiation type or radar type instrument for non contact application
what do you mean by smart instruments?explain with example
There is a problem with our HART475 it is not communicating with smart devices error becomes "NO DEVICE FOUND"we have insert 250ohm resistance as per requirement but invain plz tell me the solution?
what is loop folder?
1 Answers Bapco Bahrain Petroleum Company, Hyundai,
What is exact difference between the panel earthing and system earthing in DCS system. What is the value of resistence.
Why integral control mode call Reset controller?
In which principles tx ,pressure gauge,Temp gauge,control valve works?
0 Answers Bajaj, Professional Instruments,
what is the difference between auto tuning and manual tuning of controllers(eg:PID,PI,PD CONTROLLERS )
how can leve transmetter calibate in feild .and also put cv and flow plz .....
How can we loop powered for a transmitter.How we can do wiring
I have a close vessel tank which is under vaccum when in running condition so pl. Provide it's range for dp type tx whose flange to flange distance is 440 mm and density is 1
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)