1)why use benzophenon,uracil in hplc calibration.
2)why use sucrose in polarimeter calibration.
3)what is S/N ratio, noise and drift in hplc
4)how we can find lod and loq in method validation



1)why use benzophenon,uracil in hplc calibration. 2)why use sucrose in polarimeter calibration. 3)..

Answer / sri krishna

1. URACIL & BENZOPHENONE are commonly available standards,
with simple mobile phase i.e acetonitrile & water they can
be separated, peak shape is good, retension time is very low.
2. Sucrose is optically active which shows linearity1. Very
Easily prepared, Has a fixed rotation value which can
calibrated with 10%, 20% and 30% sugar value, and remains
constant for those concentration for a little time only,
upto that we can have a very sharp reading calibration of
pure sugar, Mainly polarimeter is used sucrose quality
detection for calibration with pure sucrose is the best
deal, Its shows almost no Muta rotation,And last and
final...Its most cheap to figure out and get the calibration
done.
3. signal to noise ratio. where signal is the response of
the analyte and noise is the disturbance present in the base
line.drift is the slow increase/ decrease the base line.
4.LOD: BY 7.1. Based on Visual Evaluation
Visual evaluation may be used for non-instrumental methods
but may also be used with instrumental methods.
The quantitation limit is generally determined by the
analysis of samples with known concentrations of analyte and
by establishing the minimum level at which the analyte can
be quantified with acceptable accuracy and precision.
7.2. Based on Signal-to-Noise Approach
This approach can only be applied to analytical procedures
that exhibit baseline noise.
Determination of the signal-to-noise ratio is performed by
comparing measured signals from samples with known low
concentrations of analyte with those of blank samples and by
establishing the minimum concentration at which the analyte
can be reliably quantified. A typical signal-to-noise ratio
is 10:1.
7.3. Based on the Standard Deviation of the Response and the
Slope
The quantitation limit (QL) may be expressed as:
QL = 10 σ/S

where σ = the standard deviation of the response
S = the slope of the calibration curve


Is This Answer Correct ?    14 Yes 1 No

Post New Answer

More Analytical Chemistry Interview Questions

In Agilent HPLC calibration using Empower 3 Software, What is the criteria for GPV Baseline starting point? i.e., Either it can start from -0.00/+0.00 or any range is there like -0.05 to +0.05 or ......etc?

1 Answers   Apotex,


what is mean by 21 CFR PART? In that what is 21 ?

1 Answers  


What is difference b/t uv detector and rid

5 Answers  


Chromatographic Purity is a Qualitative analysis

2 Answers   Shasun Chemicals,


in which situation ion pair agent require to use?

0 Answers  






in determination of(acetone) moister content we are using methanol replaced by pyredine.what hapend in that reaction.

2 Answers   Glenmark, Neuland,


How you can calculate pKa ?

6 Answers   Dr Reddys,


Difference between uv detector nd RI DETECTOR? wt s the difference between OVI Solvents nd residual solvents incase of GAS CHROMATOGRAPHY? nd wt s its detection limit?

3 Answers  


what is a drug?

11 Answers   Aurobindo,


When you get Moisture content by KF higher than by LOD, what does this indicate?

7 Answers  


what is method development parameter in HPLC method development

6 Answers   GVK,


in OSD forms require to use gas chromatography?

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)