how u meassure room temp?is it possible 2 meassure room
temp by rtd or thermocouple?between rtd n thrmocouple which
is more regard?3 meassure diff. between rtd n thermocouple?
Answers were Sorted based on User's Feedback
Answer / ramraj beda
rtd is low range temp.sensor(0-300) but thermocouple is for high range temp.(0-1400)
rtd measures temp.with change in resistance, thermocouple measures temp.with reference of cold junction, its output in mV.
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / mahender mohal
Noramaly room temp. is msrd by thermometers. yes,from RTD
we can also measures room temp. if all other perameters are
neglected but not by thermocouple because of two juctions.
Diff:- 1)rtd is low range temp.sensorbut thermocouple is
for high range temp.
2)rtd measures temp.with change in resistance, thermocouple
measures temp.with reference of cold junction, its output
in mV.
3)rtd have large bulb size than thermocouple.
4)rtd have single metal while themocouple have two
dissimiler metal wires....etc.
| Is This Answer Correct ? | 6 Yes | 2 No |
Emergency Shutdown Systems/ Plant Shutdown System.
pressure switch range 0 to 2bar.i have to install from ground level to 5 meters height.what is the formula the to increse the range ? and every 1 meter how much pressure should be add ?
what is the function of 3 way manifold valve in a closed loop system?
Why even multiple power supplu r used in instrumentation system?
what is pico fuse? how this different rom othe fuses and how to locate it inside the card?
How much Resistance Show by RTD(PT100) when it is installed in process(Consider Process Fluid temp 29 Deg C)?
What is an analog input-output
10 Answers TN, Torrent Pharma,
what is IPS (instrument protective system)?
how can we calibrate RADAR LEVEL transmitter 0% to 100%
3 Answers Chevron Phillips, Honeywell,
pls suggest suitable MOC of diaphragm for pressure measurement of UREA grade solution service fluid.
Sitrans LR460 Radar level transmitter getting level out or disturbed level during moisture present in the material ie coal in the pccb bunker in CHP.Any suggestions for this problem...????
why orfice is not used for main steam flow calculation
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)