why a ph meter require KCL solution for their storing
condition?
Answer / dr. a. bera, cdl
When not in use, the glass probe tip must be kept wet at all times to avoid the pH sensing membrane dehydration and the subsequent dysfunction of the electrode.
A glass electrode alone (i.e., without combined reference electrode) is typically stored immersed in an acidic solution of around pH 3.0. In an emergency, acidified tap water can be used, but distilled or deionised water must never be used for longer-term probe storage as the relatively ionless water "sucks" ions out of the probe membrane through diffusion, which degrades it.
Combined electrodes (glass membrane + reference electrode) are better stored immersed in the bridge electrolyte (often KCl 3 M) to avoid the diffusion of the electrolyte (KCl) out of the liquid junction.
| Is This Answer Correct ? | 72 Yes | 5 No |
What is normality
Why KBr pellet is used in IR?
wt is rsd in and explain
IF HPLC CALIBRATION FAILED THEN WHAT YOU DO
why holium oxide prepared in 10% HCLO4 SOLUTION
Why we are using a Kf Factor range 4 to 6
What is the calibration procedure for Malvern MS2000 particle size analyzer and acceptance criteria?
does water content by KF will calculate only external moisture absorbed or the ligand ? (for ex copper sulphate pentahydrate)
why 2Theta value is consider in the XRPD spectra and how to consider the 2 Theta value
What is AV value in test uniformity content ,describe calculation
how can u seperate polar compounds in hplc ?
1 Answers Meditab, Piramal Healthcare,
What is difference btw UV detector & RID detector