How can we calibrate the level interface for Fisher®
FIELDVUE® DLC3000?if use water to calibrate and our liquid
SG are 0.75 and 1 ,Displacer height=14" and How much should
be set SG in trasmitter during calibrate,for URV and LRV
how shold be set
Answer / iit
Calculation # If the calibrating liquid is water:
In the DLC3000 (Need Hart to communicate)
-ON Basic set up_PV set up set PV is Level
-ON Basic set up_PV set up set Specific Gravity =1
-ON Basic set up_PV set up_PV Rang set URV=14"and LRV=0".
1. First close both the primary isolation valves and drain
the liquid inside the chamber.
2. Fill water up to the center of lower flange level =0"
=0%
2.1 At DLC3000 on Basi set up_Sensor calibrate_Two Liquid
Lvl cal select Lower cal to get 0" =0% output
3. Fill it up to the center of the top flange level =14"
=100%
3.1 At DLC3000 on Basi set up_Sensor calibrate_Two Liquid
Lvl Cal select Upper cal to get 14" =100% output.
4. Fill it up to 50 %, check linearity.
After calibration completed with water and back on service
for interface
On DLC3000
-ON Basic set up_PV set up set PV is Interface
-ON Basic set up_PV set up set upper SG is 0.75,Lower SH=1.0
-ON Basic set up_PV set up_PV Rang set URV=14"and LRV=0".
and close drain valve and open both the primary isolation
valves unit ready to in service
| Is This Answer Correct ? | 8 Yes | 5 No |
Coriolis flow meter coil supply is 24vdc but why still we given 230ac for this transmitter let me explained detailed review please
What is voltage regulator?
I am working in a process industry, in that there are several sections devided in the SCADA screens and individual sections may START independently. But sometimes i am facing problem i.e, if section 2 Stopped by the operator immediately Section 1 or Section 3 or Section 4 is getting stopped can you expain the reason behind this?
What is the difference bitween control net scaner and device net scaner
plz you help about Que paper of mahagenco previous your,any model paper relad to it{maharastra power genration}
How will you calibrate an interface level trol/normal level trol in line and in w/s?
what is preventive maintenance , explain with difination
1 Answers Bapco Bahrain Petroleum Company,
Can a Differential Pressure Transmitter be used as a normal pressure transmitter??
Give an example where you had aproblem on the job, and tell me how you solved the problem. and the finalvoutcome.
what is cascade controller ?
what are the values of fuel flow rate in various places in boiler.
For a diafhram 1.5 d =kn(p-p0) D4 Where d in deflection in inch k=constant n=no of capdule (P-P0)=Applied pressure in psi D In active dia at capsule. -1.5 d=k to p d being constant -1.5
0 Answers L&T, Shiv Tech Marine Control,
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)