What is output of Thermocouple at 25 deg c
Answers were Sorted based on User's Feedback
Answer / b0nz0
It will depend on the type of thermocouple and the
temperature of the reference (cold) junction.
If you measure 25degC in an ambiemt temp of 25degC you
should read 0mV.
| Is This Answer Correct ? | 21 Yes | 1 No |
Answer / carl_ellis
Question: What is output of Thermocouple at 25 deg C?
I'd flunk all of you except B0nz0 because only B0nz0
understands thermocouple thermometry.
The answer is, you don't know because the cold junction
temperature is not specified. The question attempts to
find out if the concepts of a thermal gradient and cold
junction compensation are understood. Which is why, I
assume, the question does not specify the type thermocouple.
B0nz0 is correct; any type thermocouple sitting on a desk
at 25 deg C; hot end = 25 Deg C, cold end = 25 Deg C, will
read 0.0mV because there is no thermal gradient.
The cold junction temperature is the same as the hot end
temperature. Doesn't matter if it's a Type K or J or S.
With the same temperature at both ends, the output is 0mV.
But without specifying what the cold end temperature is, an
answer of 0mV assumes that the cold end is at 25 Deg C.
In the real world, it usually isn't.
The hot end can be at 25 Deg C with the cold end on a
terminal strip in control panel that's at 45 Deg C because
of the heat generated by the electronics in the control
panel. Then you'll read a negative voltage because the
cold end is hotter than the hot end!
The hot end can be at 25 Deg C and the cold end is
connected to handheld temperature calibrator in northern
Canada in the winter where the cold end terminal connection
might be at -30 Deg C.
In each case, it's the cold junction compensation that
corrects the temperature reading by measuring the
temperature of the cold end connection and taking that
temperature into account.
If I were to ask the question, I'd ask, "What is the output
of a thermocouple, both ends of which are at 25 deg C, and
why?". The answer to the first part is 0mV.
Tell 'em Carl sent you.
| Is This Answer Correct ? | 9 Yes | 0 No |
Answer / mohammed abbas
Itis depend upon the sensitivity of the thermocouple used
because the different combination of alloy is having
diffrent sensitivity like.
1,copper/contantan =0.039mv/deg
25deg c=0.915
2,Iron/contantan=0.52mv/deg
25deg c=13mv
3,Platinum/rhodium=.05mv/deg
25deg c=1.25mv
| Is This Answer Correct ? | 13 Yes | 5 No |
Answer / mohamed yousuf
it is depend upon the sensitivity of the thermocouple used
| Is This Answer Correct ? | 16 Yes | 13 No |
Answer / shabbir ahmad
first of all see it type it depends on the type of
thermocouple for example s type thermocouple has 0.113mv at
20degc every type of thermocouple has different output at
25degc.
| Is This Answer Correct ? | 1 Yes | 2 No |
what are the definitions of input, output, command & signal according to instruments controlled by PLC using DCS System?
Tyes of interface level measurement and how to use that technique for interface level measurement,
what is piezoelectric transducer?give an example in power plant n state its principle?
Why drum level range is given -500 to 0 mmwc pls give me that answer
What is Droop? What is Step/Gradient Limiter in Turbine Governor control?
How to calibrate displacer type level transmitter (chemtrol make )for level measurement in hotwell , LP & HP heater ?
units of pressure,flow, level,temperature? principle of pressure gauge? calibration of pressure tx., level tx., temperature tx.flow tx.? description of two wire ,three wire and four wire R.T.D.? WHAT IS INPUT AND OUTPUT OF TX.? WHAT IS INPUT AND OUTPUT OF R.T.D. AND THERMOCOUPLE? WHAT IS HOT LOOP AND COLD LOOP? WHAT IS INPUT AND OUT PUT AND OUTPUT OF D.C.S. AND P.L.C.? PRINCIPLE AND WORKING OF HART? WHAT IS PNP AND NPN PROXY? CALIBRATION OF PRESSURE ,LEVEL ,FLOW SWITCH? DESCRIPTION OF FIELDBUS FOUNDATION? WORKING OF SOV.?
Can anyone send most commonly asked interview question in Oil & Gas industry
Which boiler is use in power plant
How to zero trim the level transmitter by using foundation fieldbus communicator.
If fire would be inside the plant which fire detector is to detected the fire.
What is zener diode? What is voltage regulator?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)