What should be the precision for HPLC during technology
transfer?
Answers were Sorted based on User's Feedback
Answer / glp99311
The technology transfer is nothing more than reproducing
one or more method validation parameter by second
laboratory.
The instrument precision requirement is generally the
system suitability criteria (RSD) specified in the method.
The reproducibility (precision) criteria should be the same
as performed during method validation.
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / vikram
Performing Tech. transfer activity at plant location with
System precision,Method precision,LOQ precision, specificity.
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / basha
6 times of system suitability will be considered as system
precission.6 times of different assaying is method
precission
| Is This Answer Correct ? | 2 Yes | 3 No |
If compound Pka is 5 then which buffer you will select ?
how to prepare solution in ppb
why r u using hplc grade solvents
why we make titiration of Iron with EDTA ?
Why Urasil and Caffeine use for HPLC wavelength accuracy calibration. which is most suitable, Why? other than these which can be used?
At what distance the sample is collected in dissolution basket?
why the same column give different retention time in hplc?
why we use holmiumperchlorate in calibrating uv
4 Answers Aurobindo, DRL, MSN Pharma,
when we use pda detector in hplc instead of uv detector
why use benzophenone & caffene acetone in hplc calibration
Why we use caffeine for HPLC calibration in Detector wave ength accuracy test
60 Answers Arvind, Aurobindo, Cipla, Reddy Labs, Shasun Chemicals,
Why sodium hydroxide used for maintain pH of phosphate buffer