a pressure transmitter with a range of 0 to 2500kpag has a specified accuracy of
0.5 % of span.It is to be calibrated using a comparison gauge to have an accuracy
of 0.1% of span.The comparison gauge has a range of 0 to 3000kpag What is the
accuracy ratio? is this ratio acceptable?
Answers were Sorted based on User's Feedback
Answer / sreedhar
It Can Be Calibrated. The Accuracy Ratio Is 1:4 It Is Acceptable
Is This Answer Correct ? | 0 Yes | 1 No |
An atmospheric storage tank has a diameter of 20 m and a height of 8m.It is half-full of liquid with the remainder of the space containing air normally at atmospheric pressure. If the connection to atmosphere is accidentally closed off and 10 cubic m of liquid removed, what happens to the internal pressure of the tank? What are the possible consequences for the tank?
WHAT PRINCIPLE OF GAS ANALYSER( CO,CO2,N2)
different between calibration range and instrument range ?
What is the difference between Analytical instruments and other instruments?
I want to measure bidirectional steam flow? Kindly suggest me a suitable flow meter. steam is superheated at 41kg/cm2, 350°C
WHAT REASEAN USE TRANSMETER TWO WIRE & FOUR WIRE
I am using 971 Enraf radar level transmitter for LPG tanks ,the tank side indicator gives the error code 999999999FL can any body give me any diagnostic for this issue?
what is mean by hot loop and cold loop testing, whts different between both
1)how can calibrate temperature transmitter? 2)how can calibrate flow transmitter? 3)how can calibrate level transmitter?
3 Answers ccic, Emerson, JK Pharma, NIGC, Rockwell, UltraTech,
Meaning of the following a. IOP- Manimum; b. IOP- Maximu,;
when the operation engineer told you that he gives the control valve order to open but there is no response ! what will you do ? and if you found that there is astuck , what will you do ? (note : it is apanumatic valve )
What is the differerence between hysterisis and linearity.