what is meant by calibration
Answers were Sorted based on User's Feedback
Answer / sandeep sagar
the process of measuring the response of an instrument to a particular input and correcting the response to a predetermined value and repeating this in increasing and decreasing orders throughout the range of the physical parameter to be measured is called calibration.
| Is This Answer Correct ? | 25 Yes | 1 No |
Answer / mohammad reyaz ansari
Calibration is matching of the standard value of
instrumentation. Which gives Instrument society
of America (ISA) in 1968. Or matching of the value of
master (True) Instrument. 4 mA to 20 mA
3 PSI to 15 PSI, 0 % to 100 % etc.
| Is This Answer Correct ? | 3 Yes | 0 No |
Answer / parag pathak
The relationship between the physical measurement variable
input and the signal variable (output) for
a specific sensor is known as the
calibration
of the sensor. Typically, a sensor (or an entire instrument
system) is calibrated by providing a known physical input
to the system and recording the output.
| Is This Answer Correct ? | 3 Yes | 0 No |
what is maximum vaccume?
1.can you use SCADA instead of DCS,DCS instead SCADA 2.Difference between DCS and PLC,can their role be reversed i.e. role interchange 3.Difference between RTU and PLC
0 Answers IOCL, Lakshmi Machine Works,
thermocouple output is AC OR DC??
wht is loops ? how many loops are there?
Hai friends I have dought for O2 analyser how to working in boiler area. And what is the basic principles
What is the use of +/- 20ma in analog measurement,and which instrument gives +/-20ma output
What is limitation of the ht.of the displacer of a level trol?
I am a instrument technician. recently i attend interview and face one of question for DP transmitter ,the client asking about the calibration of DP transmitter ,am answer the calibration procedure, they will ask one more question. why open LP side in atmosphere? i did not answer that question. so please help me and explain the particular question. Dear firends waiting for your valuable replay
3 Answers Qatar Petroleums, Samsung,
Draw a group of instruments interlocked to a Pump & Motor?
how many types of sensor in instrumentation
How much should be the values and/or percentages of these given below factors for a good controller ? Steady state error Delay Rise Time Overshoot Peak Time Settling Time Damping Oscillations & vibrations thanks
how can be calibrate flow indication transmitter (rosemount) it shows excess milli amps nearly about 24 ma on full scale, but its scale was set on 20ma on full scale..???how can be simiulate..???
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)