i want use pressure transmitter for flow ,how?
Answers were Sorted based on User's Feedback
Answer / vishnu
It is not possible to measure flow using pressure transmitter.We need DP transmitter to measure flow.First we have to create a DP using ventury tube or a orifice plate.the flow rate is directly proportion to the squreroot of the Diffrential pressure.
Is This Answer Correct ? | 6 Yes | 0 No |
Answer / mokie
use differential pressure transmitter with oriface, vcone or
the other means
Is This Answer Correct ? | 8 Yes | 3 No |
Answer / sudarsun
Differential Pressure transducer along with a Venturi pipe
should do the work. The pressure at the normal diameter vs
the pressure at the shorter diameter should be proportional
to the rate at which the fluid flows.
Is This Answer Correct ? | 4 Yes | 0 No |
Answer / ahmed
You can't use the pressure transmitter to measure the flow
because when you want to measure the flow you need delta be
in the pressure and the presser transmitter have one side
you want transmitter with high and low sides
Is This Answer Correct ? | 4 Yes | 0 No |
Answer / jobby c johnson
In any Flow meters(venturi, Oriface etc), Flow rate is given by the expression
Q is propotional to sqrt(delta p)
Where Q is mass flow rate ; delta p is the differantial pressure across the transmitter
Is This Answer Correct ? | 4 Yes | 1 No |
How to Calibrate variable Head flowmeter (Orifice), here we dont know High pressure and low pressure then how to apply pressure on HP and LP side ?
How to work mass flow meter ?
What is the difference between Analytical instruments and other instruments?
IF I WANT TO BUY A THERMOCOUPLE THEN WITCH PARAMETER I LOOK
How many input(Analog and digital) and outputs(Analog and digital) are there in DELTA v DCS Version 7.0?
what is the difference between PLC & DCS
WHAT IS A SYMPTOM OF FAULT WITH THE SPAN OF AN INSTRUMENT? 1)ZERO-POINT AND 100% ARE CORRECT BUT NOT A MID-RANGE. 2)ZERO-POINT IS CORRECT ,BUT 1005 INPUT IS NOT GIVING 100% OUTPUT. 3)LINEARITY PRBS. 4)1005 O/p IS CORRECT BUT ZEO POINT IS WRONG.
what is interface level and how it is calibrate?
what is droop in governer ,type of droop ,how it work
What is the difference b/w EMF, voltage & potential difference?
Furnace draft transmitter range is -50 to +50 mmwc. why the impulse line connected in hp side?
In Safe Guarding test what means by "Failed to Danger" ?