In lcms analysis some times we get sodium and potassium
adducts. if the compound is havine sodium ions how can we
confirm wether sodium is from sample or it is a adduct?
Answers were Sorted based on User's Feedback
Answer / sri krishna
for sodium contamination adduct the intensity of the peak is
very low when compared to the sample sodium peak
| Is This Answer Correct ? | 3 Yes | 2 No |
Answer / spgoswami
If the precursor is sodium ion then intense peak will be of
sodium only rather than the same adduct so we can conclude
by observing lc-ms /TIC chromtgram
Regards
| Is This Answer Correct ? | 0 Yes | 1 No |