diff between dead wt calibrator n pressue comparator?
Answers were Sorted based on User's Feedback
Answer / vidal r. ramoso
ok, deadweight calibrator is an intrument with hydraulic
pressure having lifting piston with plat round platform use
to carry plate with round plat massess with standard value
use as a standard reference to clibrate pressure gauge or
any other pressurize instrsument. while the pressure
comparator is having standard calibrated pressure/vacuum
gauge use as a reference to check pressure/vacuum gauge.
pneumatic transmitter, etc.not to calibrate
| Is This Answer Correct ? | 14 Yes | 1 No |
Answer / rajendra shaha
Dead weight calibrator is secondary standard instrument use
in lab ,workshop only for calibration purpose
pressure comparator is only use for comparison of two pressure
| Is This Answer Correct ? | 5 Yes | 2 No |
Answer / maya ramesh
Answers 1 & 2 are Almost correct. But the Statement "Comparators are not used for calibrating the gauge" is not corect. If the standard test gauge used for comparing is a master test gague or having the traceiablility...Pr. Comparators are also can be used for the Calibration. In fact any home made pressure generating source also can be used for the calibration if we use the Master Test Gauge (certified) as reference.
For you information Pressure generating sources dont need any certifications or Traceiabilty.
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / rajesh
dead wt calibrator is used to measure weight / calibrate
pressure measurement
pressure comparator measures applied pressure with set
pressure (dp transmitter )
| Is This Answer Correct ? | 0 Yes | 2 No |
Where do we consider the pressure in mbar & where do we consider the pressure in bar & why is it so ?
2 Answers BOC, Jindal Steel and Power,
what is sequence table in DCS . make a logic in sequence table for tripping of ID fan in Boiler(in YOKOGAWA CS 3000 DCS).
Can anyone say that formula of Converting the Diff. Pressure into Flow in Kg/Cm2 or any explaniation of Orifice Plate, Beta Ratio , How to calculate
a laways iface problem with brine level transmitter sealing diaphram a laways giving error and maximum reading i made re range and adjust with side glass but during commission it giving gap reading between it and side glass i need help from expert to explain how to calibrate it. note:dp transmitter yokogawa company
what is difference between OP AMP and differential amplifier?
In 16' pipeline one pcv is installed which is controlled by PT installed at inlet of PCV.If Flowrate in pipeline increases what will happen to pressure at the inlet of pcv.
What is safety barrier and how it is work
can you tell me how many stages in steam turbine of 10 MW power plant
Why use ball valves on Gas pipelines instead of gate & or globe valves...means basic diffrence or use of these valves..how designers knows this type of valve with rating is required...
0 Answers Adgas, GAIL, Oil India, ONGC,
what are the standards used for earthing instruments in the field?
when i installed Allen_bradley_RsLogix_500_Pro_V_7.0,after installation it gives the fatal error.it need activation file.from where and how i can find its activation file.
What is the use of gauge?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)