IN pH CALIBRATION,CALIBRATION IS DONE FOR pH METER OR
ELECTRODE?HOW WILL YOU JUDGE THAT THE METER TO BE
CALIBRATED WITH 4,7,9.2 BUFFERS?EXPLAIN TEMPERATURE EFFECTS
ON pH.
Answer / prabhu
Hi dear the calibration is also one of the system
suitability test,whether the system working satisfactory or
not physically so when you do the calibration for pH meter
with standard pH buffer solution (which is from NIST-it may
accurate solution for actual pH)and once if your instrument
has calibrated,you can cross verify your insrument against
those pH buffer solution (NIST) and the acceptable range is
+/- 0.02 then you can conclude for ?
once if temperature raises the pH will decrease and its not
all cases
Is This Answer Correct ? | 18 Yes | 5 No |
In HPLC what is Single point threshold, peak purity index and minimum peak purity index? what should be consider for pure peak.
why we are taking 89ml concentrated HCL for preparing 1N HCL
what is the procedure for cleaning of lenses of hatr accesory of ftir instrument?
Is it necessary to inject three replicate for accuracy in AMV.
if v use HELIUM & ARGON instead of NITROGEN as a mobile phase what happen in GC SYSTEM ?
In method developement or in any analysis if we use small length column for analysis what happen then ? why ?
in dissolution if one bowl got 70 percent 2nd bowl got 80 percent and 3rd bowl got 90 percent then how proceed?
how to calibrate hplc & gc
What is rs test why we are performing rs test?
what is the meaning of ventillation in HPLC?
3 Answers Amneal Pharmaceuticals,
why 2Theta value is consider in the XRPD spectra and how to consider the 2 Theta value
what is least count of a balance? How we find out this? Any formula for least count of Analytical balance?
13 Answers Acme Formulation, Analytical Solutions, Dr Reddys, Reddy Labs,