Answer / carl_ellis

Disconnect the field sensor.

Connect a calibrated/traceable temperature signal calibrator.

Apply LRV value.

Trim the 4.0mA output.

Apply the URV value.

Trim the 20.0mA output.

Check as many interim points as the calibration protocol requires and note the output and the deviation from the expected value.

Determine whether the transmitter is within spec from the deviation values seen.

If transmitter passes, reconnect the field sensor to the input.

Fill in the paperwork.

Carl Ellis
Measure First

Is This Answer Correct ?    9 Yes 3 No

how to calibrate TT transmitter..

Answer / eid elatawy

First i have to know the connected element
1. RTD :using resistance box and divide the range into 4 quarters for example the range 0 to 100 deg we have 0,25,50,75,100 will find the related resistance from the table and see the transmitter output and if any deviation will make trim zero and span trim.
2. TC the same idea but we use MV supply and by using TC tables.

Is This Answer Correct ?    4 Yes 0 No

how to calibrate TT transmitter..

Answer / karishma

First do span and zero adjustment of a temperature transmitter. Then when the input signal is zero the output is 4mA if not readjust the span and zero again provide 25% 50% 75% 100% Signal and see the output it should give 20 mA for 100% signal input. 

Is This Answer Correct ?    1 Yes 0 No

Post New Answer

More Instrumentation Interview Questions

In your own words, would you expect an ultrasonic level device to work better in a pressure vessel with a high pressure or a strong vacuum, and why?

1 Answers  


what's the minimum level required of SIL (safety integrity level)to use fire and gas system?

2 Answers   L&T,


what is reason behind boiler drum level transmitter range put -450 to 0 mmh2o.

1 Answers   JP Group,


I need circuit for RTD senser?

0 Answers  


How much Resistance Show by RTD(PT100) when it is installed in process(Consider Process Fluid temp 29 Deg C)?

5 Answers  






what is the difference between dcs & plc

2 Answers  


sir, give me placement papers and interview quistion of birla group company of copper.

0 Answers   Birla,


what is the difference between PLC & DCS

17 Answers   L&T,


Sir, I want junior executive, Instrumentation question papaer for IOCL, Plz send to arabinda.ghosh83@gmail.com

0 Answers   IOCL,


Explain the working of rota meter.

0 Answers  


what is difference b/w the solinoid volve and autodrain?

3 Answers  


What's the difference between cold loop test and hot loop test?

5 Answers  


Categories
  • Civil Engineering Interview Questions Civil Engineering (5085)
  • Mechanical Engineering Interview Questions Mechanical Engineering (4451)
  • Electrical Engineering Interview Questions Electrical Engineering (16632)
  • Electronics Communications Interview Questions Electronics Communications (3918)
  • Chemical Engineering Interview Questions Chemical Engineering (1095)
  • Aeronautical Engineering Interview Questions Aeronautical Engineering (239)
  • Bio Engineering Interview Questions Bio Engineering (96)
  • Metallurgy Interview Questions Metallurgy (361)
  • Industrial Engineering Interview Questions Industrial Engineering (259)
  • Instrumentation Interview Questions Instrumentation (3014)
  • Automobile Engineering Interview Questions Automobile Engineering (332)
  • Mechatronics Engineering Interview Questions Mechatronics Engineering (97)
  • Marine Engineering Interview Questions Marine Engineering (124)
  • Power Plant Engineering Interview Questions Power Plant Engineering (172)
  • Textile Engineering Interview Questions Textile Engineering (575)
  • Production Engineering Interview Questions Production Engineering (25)
  • Satellite Systems Engineering Interview Questions Satellite Systems Engineering (106)
  • Engineering AllOther Interview Questions Engineering AllOther (1379)