Answer / b0nz0
It will depend on the type of thermocouple and the
temperature of the reference (cold) junction.
If you measure 25degC in an ambiemt temp of 25degC you
should read 0mV.
| Is This Answer Correct ? | 21 Yes | 1 No |
Answer / carl_ellis
Question: What is output of Thermocouple at 25 deg C?
I'd flunk all of you except B0nz0 because only B0nz0
understands thermocouple thermometry.
The answer is, you don't know because the cold junction
temperature is not specified. The question attempts to
find out if the concepts of a thermal gradient and cold
junction compensation are understood. Which is why, I
assume, the question does not specify the type thermocouple.
B0nz0 is correct; any type thermocouple sitting on a desk
at 25 deg C; hot end = 25 Deg C, cold end = 25 Deg C, will
read 0.0mV because there is no thermal gradient.
The cold junction temperature is the same as the hot end
temperature. Doesn't matter if it's a Type K or J or S.
With the same temperature at both ends, the output is 0mV.
But without specifying what the cold end temperature is, an
answer of 0mV assumes that the cold end is at 25 Deg C.
In the real world, it usually isn't.
The hot end can be at 25 Deg C with the cold end on a
terminal strip in control panel that's at 45 Deg C because
of the heat generated by the electronics in the control
panel. Then you'll read a negative voltage because the
cold end is hotter than the hot end!
The hot end can be at 25 Deg C and the cold end is
connected to handheld temperature calibrator in northern
Canada in the winter where the cold end terminal connection
might be at -30 Deg C.
In each case, it's the cold junction compensation that
corrects the temperature reading by measuring the
temperature of the cold end connection and taking that
temperature into account.
If I were to ask the question, I'd ask, "What is the output
of a thermocouple, both ends of which are at 25 deg C, and
why?". The answer to the first part is 0mV.
Tell 'em Carl sent you.
| Is This Answer Correct ? | 9 Yes | 0 No |
Answer / mohammed abbas
Itis depend upon the sensitivity of the thermocouple used
because the different combination of alloy is having
diffrent sensitivity like.
1,copper/contantan =0.039mv/deg
25deg c=0.915
2,Iron/contantan=0.52mv/deg
25deg c=13mv
3,Platinum/rhodium=.05mv/deg
25deg c=1.25mv
| Is This Answer Correct ? | 13 Yes | 5 No |
Answer / mohamed yousuf
it is depend upon the sensitivity of the thermocouple used
| Is This Answer Correct ? | 16 Yes | 13 No |
Answer / shabbir ahmad
first of all see it type it depends on the type of
thermocouple for example s type thermocouple has 0.113mv at
20degc every type of thermocouple has different output at
25degc.
| Is This Answer Correct ? | 1 Yes | 2 No |
Hi, am aniket, I got engineering degree in instrumentation and control, but as their were no job availabilities for engineer's post I started working as instrumentation technician I kuwait, so my question is is this technician's experience worth for getting the position of engineer?
1 Answers Almeer Technical Services,
Explain about SOURCING & SINKING concepts in PLC to field sensor wiring ?
why we use negative 24vdc to a proximeter
what is the process to make control schematic ?
I am applied 100 % pressure on to the air to open control valve,but the valve is not fully closed..and 10-15% is open the valve. how to calibrate the control valve?
What skill/s does the Instrument Technician has?
We have ZR22G-100-S-A-E-T-T--E-A/CV/F1 Model Zirconia Oxygen analyser which is Yokogawa make. I kept the analyser in Flue gas path at Air pre heater area. My plant is 7.5MW Biomass power plant. For a particular fuel, Coconut Husk which contains Sodium and Potassium chlorides iam not getting proper reading, Analyser output always Zero. I calibrated the analyser and it is OK. For Other fuels like Wood chips, Cholam waste, DOB, Groundnutshells, Ply wood waste and all iam getting actual percentage of Oxygen. What is the main reason behind this draw back?
draw shematic diagram of c/v positioner(smart type)?
The pressure class rating (such as 150 lb, 300 lb, etc.) of a line appears in the form of ________________ in the Line Number (or Line Designation) on a P&ID
is fan is a transducer or inverse transducer?
in which mode of the zener diode act when load voltage is approximately constant
need instrumentation job completed 1 year trainee ship in mnc pesticide company. now in search of job
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)