what r the parts calibrated in gc
Answers were Sorted based on User's Feedback
Answer / mahesh
Many parts like Auto injector, Detector,Injection volume,
Flow rate,
Is This Answer Correct ? | 21 Yes | 1 No |
Answer / kuchela rao athota
1)CARRIER FLOW CHECK
2)COLUMN OVEN TEMP..(ISOTHERMAL TEMP.. CHECK)
3)COLUMN OVEN TEMP..(PROGRAMMABLE TEMP..CHECK)
4)VIAL HEATING BLOCK TEMP..
5)TRANSFERLINE TEMP..
6)PRECISION TEST
7)LENIARITY TEST
8)FID SENSIVITY TEST
Is This Answer Correct ? | 15 Yes | 3 No |
Answer / alpesh patel
Parts like Auto injector, Detector, Oven, HS
Is This Answer Correct ? | 4 Yes | 1 No |
Answer / jyothi
gas flow rate
carrier gas flow rate
detector gas
oven calibration
detector calibration
these above all calibration parameters in gas
chromatography.
Is This Answer Correct ? | 4 Yes | 1 No |
Answer / a.ramakrishnan
Parameters for GC Callibration
1. Carrier gas flow rate.
2. Split Flow Callibration.
3. Injector Linearrity
4. Detector Response
5. Column oven temp.
Is This Answer Correct ? | 3 Yes | 0 No |
Answer / surya
Flow rate , Oven , Auto injector , Detector response , HS
Is This Answer Correct ? | 3 Yes | 2 No |
Answer / vimal
AUTO INJECTOR
DETECTOR
COLUMN OVEN
VIAL TEMPARATURE
HS
Is This Answer Correct ? | 2 Yes | 1 No |
Answer / kanakadurga
Calibration parameters in Gc
1.Flow calibration:hydrogen gas,air and carrier gas
2.Column oven temperature:50,100,200,300
3.Injector precesion:6 times
4.Detector linearity:20% to 120%
Is This Answer Correct ? | 1 Yes | 0 No |
Answer / shanthi
Injector precission
Liearity of the detector
Oven temperature calibration
Carrier gas flow rate calibration
Is This Answer Correct ? | 1 Yes | 0 No |
Answer / s v ramesh
1. OVEN TEMPERATURE
2. HSS SENSITIVITY WITH USING ETHANOL OR IPA
3. CAPILLARY SENSITIVITY OF FID USING FID SAMPLE
4. DETECTOR LINEARITY.
Is This Answer Correct ? | 0 Yes | 0 No |
IN DISSOLUTION S1 STAGE 1 BOWL GOT 52%(Q POINT IS 80%)WHAT WE HAVE DO
How to start the dissolution development for unknown tab?
What is difference b/w IR&NIR
in loss on drying Why follows on constant weight not more than 0.5 mg, any region
WHICH STANDARD WE USE IN SILICA ANALYSIS IN SILICA-GEL METHOD FOR SOLID SAMPLE
why we use vanillin in melting range calibration
why the same column give different retention time in hplc?
If suppose 10 methods of dissolution given in pharmacopoeial for single content product then which method out of 10? or all 10 require to follow?
pda detector mechanism?
1 Answers DKR, Teva Pharmaceuticals,
IF HPLC CALIBRATION FAILED THEN WHAT YOU DO
What is cleaning validation of rinse and swabe and its limit with caculation formula.
how pyridine can be analysed other than GC?