gc Calibration % RSD failed what is the possible reasons?
Answers were Sorted based on User's Feedback
Answer / hiren
First is not column proper installations
Secound is not proper work syrring
Thred is injector septa so much punching
| Is This Answer Correct ? | 22 Yes | 0 No |
Answer / yogesh patel
flow rate of gas not proper its cose fluctuation in assay
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / alasyam ramesh
1) Improper stabilisation of column and system.
2) injector septa bleeding
3) partial blockage of detector jet.
4) syring struck due to solvents nature( solvent viscosity).
5) plunger moving and taking improper injection volume.
6) some time may effect room temperature also
| Is This Answer Correct ? | 5 Yes | 0 No |
If %RSD is failing in GC calibration means we have to check & replace following things,
1.Injector septa
2.glass wools in liner
3.syringe
Thanks,
B.Gangatharan.
| Is This Answer Correct ? | 4 Yes | 0 No |
why you require QA approval & what is its importance?
How many clauses in ISO!
In my analysis by HPLC,In sample main peak and placebo peak are eluting with shorter resolution ( almost look like doublet). how can i get max resolution ? Is changing gradient method is usefull?
Weight balance variation limit during analysis
7 Answers Amneal Pharmaceuticals,
if content uniformity passing but dissolution varrying then what is next step?
Which Zone of our country like conducting of Stability studies(Accelerated & Long term stability studies)as per ICH guidelines.
what you do when compound having dwel wavelength?
What is Endcapped Hplc column
Why red HgI2 added during starch preparation?
for which type of applications gas chromatography is used?
how to prepare 5normality of hcl
In chromatograpbic purity 90% is there, in assay 70% is there. How will you prove remaining 30%.