Answer / sri krishna
Condensation Test
Use this test whenever injector or carrier gas contamination
problems are suspected (e.g., ghost peaks or erratic baselines).
* Leave the GC at 40-50�C for 8 or more hours.
* Run a blank analysis (i.e., start the GC, but with no
injection) using the normal temperature conditions and
instrument settings.
* Collect the chromatogram for this blank run.
* Immediately repeat the blank run as soon as the first
one is completed. Do not allow more than 5 minutes to elapse
before starting the second blank run.
* Collect the chromatogram for the second blank run and
compare it to the first chromatogram.
* If the FIRST chromatogram contains a substantially
larger amount of peaks and baseline instability, then that
is an indication that there is contamination upstream of the
capillary column (ie. contaminated inlet, dirty carrier gas,
etc.).
* If BOTH chromatograms contain few peaks or very little
baseline drift, it can be assumed that the carrier gas
and/or inlet are relatively clean.
* If BOTH chromatograms contain a significant amount of
noise and/or baseline drift, then that usually is an
indication that the detector or detector gases are contaminated.
| Is This Answer Correct ? | 8 Yes | 1 No |
Answer / guest
retention time per puss we r using mobile phase ph
| Is This Answer Correct ? | 0 Yes | 0 No |
why has inject 5 injection in assay
how a particular wavelength can be different for a particular compund while analysing by uv and by HPLC.
What is the difference between STANDARDIZATION and CALIBRATION of instrument like KF? why do we standardization of KF everyday? Why the KF factor being only 5.02 etc?
5 Answers AlpaMed, Vergo pharma research,
Why pyrene is also use for calibration if detector
in which situation require to change rs specification?
using gradient pressure in gas chromatography are not ?using gradient pressure why
what is stability study ? why it is needed and how itis done
Why do we use sodium salicylate for Calibration TLC Cabinet U.V Lamps
In Agilent HPLC calibration using Empower 3 Software, What is the criteria for GPV Baseline starting point? i.e., Either it can start from -0.00/+0.00 or any range is there like -0.05 to +0.05 or ......etc?
How to establish relative response factor for hplc. Why it is required?
How to set specification of assay, fisdolution and related substances?
why kf factor coming 0.5