Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


In GC, What does 'Column condensation test' mean by?

Answers were Sorted based on User's Feedback



In GC, What does 'Column condensation test' mean by?..

Answer / sri krishna

Condensation Test
Use this test whenever injector or carrier gas contamination
problems are suspected (e.g., ghost peaks or erratic baselines).

* Leave the GC at 40-50&#65533;C for 8 or more hours.
* Run a blank analysis (i.e., start the GC, but with no
injection) using the normal temperature conditions and
instrument settings.
* Collect the chromatogram for this blank run.
* Immediately repeat the blank run as soon as the first
one is completed. Do not allow more than 5 minutes to elapse
before starting the second blank run.
* Collect the chromatogram for the second blank run and
compare it to the first chromatogram.
* If the FIRST chromatogram contains a substantially
larger amount of peaks and baseline instability, then that
is an indication that there is contamination upstream of the
capillary column (ie. contaminated inlet, dirty carrier gas,
etc.).
* If BOTH chromatograms contain few peaks or very little
baseline drift, it can be assumed that the carrier gas
and/or inlet are relatively clean.
* If BOTH chromatograms contain a significant amount of
noise and/or baseline drift, then that usually is an
indication that the detector or detector gases are contaminated.

Is This Answer Correct ?    8 Yes 1 No

In GC, What does 'Column condensation test' mean by?..

Answer / guest

retention time per puss we r using mobile phase ph

Is This Answer Correct ?    0 Yes 0 No

Post New Answer

More Analytical Chemistry Interview Questions

why has inject 5 injection in assay

2 Answers   Lupin,


how a particular wavelength can be different for a particular compund while analysing by uv and by HPLC.

0 Answers  


What is the difference between STANDARDIZATION and CALIBRATION of instrument like KF? why do we standardization of KF everyday? Why the KF factor being only 5.02 etc? 

5 Answers   AlpaMed, Vergo pharma research,


Why pyrene is also use for calibration if detector

1 Answers   Cipla, Dr Reddys,


in which situation require to change rs specification?

0 Answers  


using gradient pressure in gas chromatography are not ?using gradient pressure why

0 Answers   Reliance,


what is stability study ? why it is needed and how itis done

2 Answers   Dr Reddys, RB,


Why do we use sodium salicylate for Calibration TLC Cabinet U.V Lamps

3 Answers   Matrix, Spar Labs,


In Agilent HPLC calibration using Empower 3 Software, What is the criteria for GPV Baseline starting point? i.e., Either it can start from -0.00/+0.00 or any range is there like -0.05 to +0.05 or ......etc?

1 Answers   Apotex,


How to establish relative response factor for hplc. Why it is required?

0 Answers  


How to set specification of assay, fisdolution and related substances?

0 Answers  


why kf factor coming 0.5

5 Answers   Alkem Labs,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)