Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
Any molecule is filing for ANDA purpose, what are the dissolution requirements / specifications needed ?
what is the results if we use ODScolumn instead of BDS column ,(ther parameter is same for column) in assay method ?
Why caffeine as a reference standard used for calibration of HPLC?
11 Answers Gangwal Chemicals, Orchid,
what is the difference between system precision and method precision?
3 Answers Teva Pharmaceuticals,
what is mean by assay?
What is the major difference b/w Drift and Noise
4 Answers Aurobindo, Dr Reddys, Reddy Labs,
Which column is more polar, C8 or C18?
6 Answers Cipla, Glenmark, Laurus, Torrent Power,
What is the pH of water?
what is the difference between isocratic & elution technique?
What is the difference between bulk property detectors and Solute property detectors vivimed labs limited
Describe your experience with analytical instrumentation?
3 Answers Alembic, Apotex, Aurobindo, DLF, Enal Drugs, Lupin, Natco, Pharma,
What is a difference between potency and purity?