How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
Is This Answer Correct ? | 8 Yes | 8 No |
How to compare XRD graphs against standard and carry polymorpism study of API's by powder XRD method?
WHAT IS USE OF ION PAIR REAGENTS?
is it necessary to do HPLC calibration at wavelength 315nm if we are doing analysis at this wavelangth
Shall i use a same hplc column for two different products.?
what is the difference between EP and BP
2 Answers Novartis, Vimta Labs,
Why sodium hydroxide used for maintain pH of phosphate buffer
How to do spike of aflatoxin in peanut to get maximum recovery
what is the dipolement and how it is related to IR
1 Answers AJ Square Inc, Aurobindo,
How can we calculate "confidence interval" in analytical method validation? Pl. explain with example.
what is the difference between Tailing Factor & Asymmetry
8 Answers Ashish Life Science, BEC Chemicals, Jubilant,
In Karl fischer titrator what medium we can choose to analysing aldehyde and ketone samples by using KF reagent
What are modes of quantification in related substances