How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
Why sodium lamp is used as source of Polarimeter?
12 Answers College School Exams Tests, Hetero, NCL, Zydus Cadila,
How we develop a method by HPLC for unknown compound,if we got three peaks with same response then how we can find out which one is our desired compound's peak without taking help of LC MS.
what is mean by covalidation
Many times I don't got a caffeine peak in calibration of hplc using guard column ❓
why we are doing seal wash and needle wash?
what is partion and column chromatography
Which type of column should i use to check the purity of high molecular weight protein using HPLC reverse phase column chromatography? Hi everyone. I wanna to check the purity of high molecular weight protein (collagen) with MW of ~130 kDa using a HPLC. I know C18
0 Answers Petronas, Shanghai Ocean University,
effect of pore size, pore volume, partical size, column length, carbon load on retention time? what is carbon load? what is the use?
How to compare XRD graphs against standard and carry polymorpism study of API's by powder XRD method?
is there any material other than caffene to calibrate hplc detector
How we performed the force degradation for drug substance, is any specific guideline is available for each parameter(Acidic, basic, oxidation,heat)? what conditions you mentained for above parameters.
What is aggregate and fragments in SEC?