What is the difference between low pressure gredient and high
pressure gredient in hplc instrument?

Answers were Sorted based on User's Feedback



What is the difference between low pressure gredient and high pressure gredient in hplc instrument..

Answer / yojana

In a low pressure gradient pump, different mobile phase is
mixed using a valve before entering the pump head. In a
high pressure gradient pump, each different mobile phase is
delivered by an individual pump head and then mixed at the
pump outlet. The low-pressure gradient design is of lower
cost as it only use one pump head. It can also use more
types of mobile phase without significant increase of cost.
The high-pressure gradient design provides a much faster
gradient since solvent mixing point is much closer to the
column head. This is measured using delay volume. The value
can be 50-300 uL for high pressure gradient pump and can be
2 to 3 times larger for a low pressure gradient pump. A
small delay volume is important when the analysis time is
short or the flow rate is low. If the delay volume is too
large, it become impossible to obtain reproducible gradient
run since the planed composition cannot reach the column
head before a run is finished..

Is This Answer Correct ?    22 Yes 4 No

What is the difference between low pressure gredient and high pressure gredient in hplc instrument..

Answer / b.gangatharan

In simple definition,we can say that in low pressure gradient sample mixing with mobile phase will take place before pump and in high pressure gradient it will take place after pump.
Thanks,
B.Gangatharan,
Caplin point Ltd-unit-1
Pondicherry.

Is This Answer Correct ?    2 Yes 0 No

What is the difference between low pressure gredient and high pressure gredient in hplc instrument..

Answer / analyticalstudent

Is This Answer Correct ?    0 Yes 1 No

What is the difference between low pressure gredient and high pressure gredient in hplc instrument..

Answer / srinivask1985

In Low pressure gradient we are using single pump and one
more resorvoirs.
In High pressure gradient we are using one more pumps and
one more resorvoirs.

Is This Answer Correct ?    4 Yes 7 No

Post New Answer

More Analytical Chemistry Interview Questions

Can we do the MC analysis of a sample which is not completely soluble in the medium choosed.

2 Answers  


What is colimn selection Criteria in HPLC analysis?

4 Answers  


how to calibrate hplc & gc

0 Answers   Cipla,


why buffer is added in solvent system in HPLC ??

10 Answers   Lupin,


what is peak purity? how may types of methods to determine peak purity?

3 Answers   Mylan,






how to calculate elemental impurities?

0 Answers  


In HPLC (Waters) software Empower1,2 &3 what is the updation for each version

0 Answers   APL,


Why base line not start from zero in GC Chromatography analysis.

1 Answers   Orchid,


Why dst is used in kf for factor determination less than 1

4 Answers   Reddy Labs, Sun Pharma,


WHAT IS THE DIFFERENT BETWEEN PURITY, POTENCY AND ASSAY

6 Answers   Glochem, Hetero,


what is the cleaning validation?

1 Answers   BioPlus,


about method developement in hplc

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)