Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


What is the procedure for investigation of an OOS (Out of
specification) results?

Answers were Sorted based on User's Feedback



What is the procedure for investigation of an OOS (Out of specification) results?..

Answer / b.suresh naidu

OOS may be due to assignable cause (or) Non-assignable
cause.

Assignable cause: It means error is identified.
Non-assignable cause: It means error is not identified.

When any OOS result is observed, first to know any
assignable cause Laboratory preliminary investigation
(Phase-I) is recommended. Phase I investigation should
contain whether correct methododology followed as per STP,
used calibrated instruments,Analyst trained & re-injection,
re-filtration, re-dilution, re-sonication from the
original stock solutions, _ _ _etc).

If any Laboratory error is identified then repeat analysis
as defined in the SOP [Min it should be NLT 6 replicates by
two different analysts (preferably Original analyst &
experienced analyst)]

If no Laboratory error is identified, then full scale
investigation (Phase-II) is recommended. Phase-II
investigation should contain atleast Manufacturing,
Packaging, sampling/re-sampling, _ _ _etc.This sholud be
clearly defined in the SOP.

If any assignable cause is identified then repeat analysis
as defined in the SOP [Min it should be NLT 6 replicates by
two different analysts (preferably Original analyst &
experienced analyst)].If no assignable cause is identified
in phase-II investigation, then further investigation/
batch disposition shall be done as recommended by QA.

Further more details refer CDER guidelines from USFDA.

Is This Answer Correct ?    36 Yes 1 No

What is the procedure for investigation of an OOS (Out of specification) results?..

Answer / rams

OOS Investigation done by groupleader ,firstly wether check
analyst follow the correct STP ,check the analysis he is
done e.g. he used proper cleaned glassware and correct
working standard ,he used calibrated system like Ph
meter,Dissolution,HPLC and also correct dilutions .If he
done correct analysis,then give same sample for 2nd analyst
for OOS analysis.

Is This Answer Correct ?    16 Yes 5 No

What is the procedure for investigation of an OOS (Out of specification) results?..

Answer / ramya

when oos arises inform to the supervisor and discuss the analyst procedure whether they followed wrong or right.if right then repeat the analysis by different analysts.then also result is same means inform to manager and QA DEPT.Get approval from QA and fill the form with obtained results.

Is This Answer Correct ?    2 Yes 3 No

What is the procedure for investigation of an OOS (Out of specification) results?..

Answer / santosh

Refer CDFR gideline from USFDA

Is This Answer Correct ?    5 Yes 8 No

Post New Answer

More Analytical Chemistry Interview Questions

is a bonded water molecule in an organic material can be easlisy and accuractly determined by KF rather than LOD

2 Answers  


how calibrate the hplc detector

2 Answers   Cipla, Shraddha Analytical Services,


What is oot in Stability analysis? (example)

2 Answers   Indoco Remedies,


When we r doing processing method for sst in Empower software by HPLC, What value has to give in void volume and as per which pharmacopia!

5 Answers   Micro Labs, Mylan,


why should we calibrate pH meter with 4,9.2 and 7.4 buffers particularly?

5 Answers   Celon Labs,


IF WE INCREASE FLOW RATE RT DECRESE.WHY

2 Answers  


what is the difference between capillary and glass column in GC

1 Answers  


I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?

0 Answers  


What is heavy water

3 Answers   Meghmani Organics,


How to choose the which salt is suitable for mobilephase

0 Answers   Cipla, Ranbaxy,


How is the Flame test performed?

4 Answers   Lupin,


which batch require to use for analytical method validation?

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)