How to calibrate glassware against std calibrated glassware?
Answers were Sorted based on User's Feedback
Answer / ramalingeswararao
APPLYING THE WATER FACTOR AT 25 DEG CENTIGRADE 0.99608.
Is This Answer Correct ? | 1 Yes | 0 No |
Answer / jitender axiva
Calibration procedure is nothing but comparision of one of
the know magnitude value erf. as std to the test one.
in the case of glassware std calibarted glassware must be
10 times (least count)higher accuracy than test unit.
e.g. if we want to calibrate 10 ml pipette then we must
have to calibrate it through pipeete which of 1 ml with
least count less than that of test one.
Is This Answer Correct ? | 2 Yes | 3 No |
Answer / mr.bujji reddy kanchi
Designated volume, mL 10 25 50 100 250 500 1000
Limit of error, mL 0.02 0.03 0.05 0.08 0.12 0.20 0.30
Limit of error, % 0.20 0.12 0.10 0.08 0.05 0.04 0.03
Weight measurement by taking water against standard
glassware(class-A type)
Is This Answer Correct ? | 2 Yes | 5 No |
What is the requirement for brazil in the analytical method vslidation comparr to ich?
Basic principle of ESI?
What is the difference between high pressure gradient and low pressure gradient in hplc?
in calibration of disintegration test apparatus why cycle limit is 28-32?
if you have given one product then how you determine the impurity in that?
what is DIPE limit in residual solvent?
Why potassium dichromate was using in Uv calibration?
Difference Between End Cap HPLC Column and Non End Cap Column
Why linearity of detector performed in HPLC calibration
Which detector is used in uv-vis spectrophotometer?
what are the important specifications for a Column ?