Answer / vipul
For Assay:- Sample concentration should be lower than higher
detector linearity concentration.
For Related substances:- Sample concentration should be
decided on the basis of LOD and LOQ determination
e.g sample concentration 1000ppm
Standard low load concentration 5ppm(For specified imp
limit-0.5%)
LOQ-should be between 0.05% To 0.09% of sample concentration.
| Is This Answer Correct ? | 8 Yes | 2 No |
Answer / pramod
It is also depend upon peak response.
How ever we have to select sample concentration in rs on the basis of LOQ
"Loq should be less than Reporting threshould"
Accordign to that we have to select sample Concentration
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / rajareddy
based on the peak response ,peak height and LOD and LOQ Signal. Ex:sample concentration 1mg/ml than LOQ value (0.1%mg/ml) LOD value(0.03mg/ml)
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / asha latha
IF ALL IMPURTITIES AT SPEC LEVEL ARE MEETING S/N RATIO MORE
THAN 60 AT THAT PARTICULAR CONCENTRATION AS WEL AS NO
COLUMN OVERLOAD.
| Is This Answer Correct ? | 0 Yes | 6 No |
What is the use of ELSD in HPLC?
Why are use silicon oil mr?
How will you separate Colorless compound Using Column Chromatography?
4 Answers Delhi University, VPHS,
Why a1% value is used for some product ? What is the criteria for selection a1% value ?
what is difference between zero air gas and o2 gas
1 Answers IPCA, Vasudha Pharma,
If compound Pka is 5 then which buffer you will select ?
If inhouse hplc related substance method is completly diffrent from Usp for finished proďuct with diffrent impurities then how require to prove method equivalecy?
What are the primary standards?
which procedure we follow if 6th month of accelerated sample fail?
why we inject sample TWICE in ASSAY by HPLC analysis but in RS once?
4 Answers Aurobindo, DRL, Neuland,
what is the difference between Hydrochloric Acid & Hydrogen chloride
What is the name of the detector used in TOC?