How do we fix the sample concentaryion in hplc method
development. what is the basis?Thanks in advance

Answers were Sorted based on User's Feedback



How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / vipul

For Assay:- Sample concentration should be lower than higher
detector linearity concentration.
For Related substances:- Sample concentration should be
decided on the basis of LOD and LOQ determination
e.g sample concentration 1000ppm
Standard low load concentration 5ppm(For specified imp
limit-0.5%)
LOQ-should be between 0.05% To 0.09% of sample concentration.

Is This Answer Correct ?    8 Yes 2 No

How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / pramod

It is also depend upon peak response.
How ever we have to select sample concentration in rs on the basis of LOQ

"Loq should be less than Reporting threshould"

Accordign to that we have to select sample Concentration

Is This Answer Correct ?    2 Yes 0 No

How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / rajareddy

based on the peak response ,peak height and LOD and LOQ Signal. Ex:sample concentration 1mg/ml than LOQ value (0.1%mg/ml) LOD value(0.03mg/ml)

Is This Answer Correct ?    0 Yes 0 No

How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / asha latha

IF ALL IMPURTITIES AT SPEC LEVEL ARE MEETING S/N RATIO MORE
THAN 60 AT THAT PARTICULAR CONCENTRATION AS WEL AS NO
COLUMN OVERLOAD.

Is This Answer Correct ?    0 Yes 6 No

Post New Answer

More Analytical Chemistry Interview Questions

what is the purge flow & how to calculate

0 Answers   Granules,


why urasil use in hplc column perfomance?

1 Answers   Bayer,


give clarity of linearity and range in method validation

0 Answers  


If compound Pka is 5 then which buffer you will select ?

5 Answers   Avra, GVK,


what is hold time in Gc?

1 Answers  






IN DISSOLUTION S1 STAGE 1 BOWL GOT 52%(Q POINT IS 80%)WHAT WE HAVE DO

9 Answers  


What is the purpose & importance of using the polystyrene film in ir calibration?

3 Answers   Vimta Labs,


Explain about Low pressure gradient and High Pressure gradient systems in HPLC?

4 Answers   Mylan, Orchid, Roshanico,


What is GF254 means in TLC plate?

1 Answers   Abbott,


What is the principle of Thermal conductivity detector and FID?

0 Answers   Cipla,


what is the maximum flow rate in prep hplc

1 Answers   Meril Life Science,


Which water can use Hplc analysis

9 Answers   BioGenomics, Dr Reddys, Reddy Labs,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)