How to Calibrate Transducers? Can you give me a step by ste
procedure?
Answer / mahesh bajaj
The transducers are calibrated against reference standards
which are atleast 10 times more accurate then the transducer
to be calibrated.
Let me elaborate by giving an example.
Orifice plate infers flow rate by measuring differential
pressure.so to calibrate orifice plate, reference standard
say magnetic flowmeter is used which measures flow rate.say
flow range is 0-100 lpm and at every 10 lpm, differential
pressure across orifice is measured and constant is evolved
between flow rate and differential pressure. This constant
is also known as discharge coefficient or calibration
constant and one dimensional bernolli's equation is used to
compute this constant.
Now once you know the constant the orifice flow element is
ready to be used in any line size without using any
reference standard.
Hope this helps.
| Is This Answer Correct ? | 7 Yes | 1 No |
What is loop checking ?
10 Answers Kharafi National, Petrofac, Samsung,
SOME OF YOUR QUESTIONS ARE NOT MAKING ANY SENSE SO TRY AND PUT IN A LITTLE BIT MORE EFFORT BEFORE YOU POST A QUESTION. WE WILL ONLY HELP THOSE WHO WE CAN SEE ARE TRYING TO HELP THEMSELVES.
why any measurement made interms of pressure is more accurate ? to be more precise most of the flow velocity is made interms of pressure and it is accurate. what is the reason behind pressure measurement being accurate ?
In apron feeder how TPH is calculated?
What is the difference between PLC and DDC?
NAME ANY TWO ELASTIC PRESURE SENSOR?
How to calibrate H2s Dectector
you must show the examples application and give at least 4 types of transducer/sensors that can minimize the failure during the manufacturing of the resistors.
i want know about capillary type LT transmitter installation. we can install any location in between to HP and LP leg?
why the supply frequency used in America is 60Hz while that used in India is 50Hz?,
any body tell me ,contol valve has how many cheractristic? explain it.
In my company plc based winders are operated. one day there was a data transmission failure alarm shown on pc. as per sir instruction we have reset master plc but winders were not triped and alarm gone. so what master plc do?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)