Answer / pankaj shrivastava
There are some factors and parameters by which we can
decrease the run time as....
1. by increasing flow Rate
2. by increasing lightly Column oven temperature.
3. by increasing mobile phase strength.
3. by using short length column.
Is This Answer Correct ? | 28 Yes | 3 No |
Answer / priyanka.r
1)By changing the mobile phase composition of organic solvent(for ex:acetonitrile),then RT will shift.
2)by changing the flow rate, length of the column, temperature.
3)the next option is better choose some advanced Technic like UPLC.
Is This Answer Correct ? | 13 Yes | 1 No |
1) By increase the mobile phase ratio of organic solvent,
2)By increasing Flow rate, 3)Go for Gradient Progrom 4)
Reduce the column Length And Particle size 5)By Changing the
pH of the mobile phase RT will shift
Is This Answer Correct ? | 5 Yes | 0 No |
Answer / nirav
By addition of organic solvent in mobile phase, By changing
the flow rate, By changing length of the column.
Is This Answer Correct ? | 3 Yes | 3 No |
Answer / ra
You must try for UPLC.Or otherwise you have to do hard work on the method.
Is This Answer Correct ? | 3 Yes | 8 No |
Answer / santhu
BY ADJUSTING THE PH OF THE MOBILE PHASE WITH BUFFERS
Is This Answer Correct ? | 1 Yes | 7 No |
What is the difference between uplc and hplc
4 Answers AAU Assam Agricultural University, Hetero,
cefoperozone and sulbactam inj. hplc test method
0 Answers Graviti Pharmaceuticals,
What is the Column Endcaping in HPLC ?
Why we use 100 RPM in friability test
acetic acid is weak acid & perchloric acid is strong acid.but how perchloric acid is soluble in acetic acid
Which is the best mobile phase to separate organic acids using C8 column. And how should the samples be prepared for hplc?
Why we are having specification 100% above in case of assay as there is only 100% of compound?
2 Answers MSN Pharma, Nectar Lifesciences,
Why should we use Potassium iodide and Hydrochloric acid during preparation of Iodine Molar solution?
What is Isotope tracer technique?
1 Answers LAP, MAHINDRA, MPSC, PSI Data Systems, WNS,
Why acetonitrile and water are used as extraction solvent when analysis melamine? I thought they are miscible and won't be able to separate...
in hplc calibration injector linearity pass,it means inj and det is good. why to perform detector linearity is it required, if it is required,what is difference between injector and detector linearity in hplc calibration
IS THERE ANYWHERE MENTION THAT THE RE PROCESS BATCH REQUIRE TO KEPT STABILITY STYDY