In case of dissolution test, if there is variation within 6
individual units however the batch is complying as per S1
criteria, is it necessary to do investigation for the same?
What should be the criteria for such variation?
Answers were Sorted based on User's Feedback
Answer / sasikumar.r
Not necessaty to investigate.....
As per USP Acceptance Criteria for
S1 is 6 Each unit is not less than Q + 5%.
If individual values passes the Q+5% means, that batch passes the dissolution..
| Is This Answer Correct ? | 3 Yes | 0 No |
Answer / b.kishan reddy
Dear @Shasikumr %RSD Should be match for the 6 Units. Generally it is 5%.
| Is This Answer Correct ? | 0 Yes | 0 No |
how to calculate corelation coefficient in G.C Calibration? what is the calculation part?
what are the different methods of analysing related substances test,what are LOD AND LOQ .HOW TO REPORT IMPURITIES WHEN LOD AND LOQ LIMITS R SPECIFIED
how to prepare 5normality of hcl
iam using ph instrument lab india make.3ponit and 5point.in calibration time 3point slope will come 98.89and 96.87.that limit is 85 to 105.what basis slope will come.any calucalation between them.5point wiicome slope 99.89?why 3p0int came 2 values
Why the same type of columns showing different RT for a single product?
what is the cleaning validation?
why IR graph or spectra comes down ? Any reason
Why 10%,20%,30% Sucrose solution is used for the calibration of polarimeter?
5 Answers Cipla, LGC Promochem,
what is the extinction coefficient for Indomethacin or at 319nm wave length.
0 Answers GSK GlaxoSmithKline,
Why chromic acid wash is not recommended by USP for glassware cleaning?
if relative response factor is 1.07, to select how much value
How will you select mobile phase in HPLC Method Development