In my analysis by HPLC,In sample main peak and placebo
peak are eluting with shorter resolution ( almost look like
doublet). how can i get max resolution ? Is changing
gradient method is usefull?
Answers were Sorted based on User's Feedback
Answer / wasim shaikh
You can try one or more of the following:
1) Changing the column type (c8 to c18, etc.)
2) Changing the Mobile phase composition (increase buffer
etc.)
3) Decrease the particle size on column (from 5 to 4, 3
etc.)
4) Increase the column length
5) Decrease the flow rate
6) If all of the above fails, then as you suggested,
gradient method can be used to separate the two peaks.
P.S. : Gradient method can also be used as an alternative
to the above methods and not necessarily the ultimate
solution.
| Is This Answer Correct ? | 32 Yes | 1 No |
Answer / naveen
U CAN DECREASE THE FLOW RATE
AND INJECTION VOLUME
| Is This Answer Correct ? | 1 Yes | 1 No |
Answer / gaurav
You can increase the injection volume.Changing gradient
method is not required.
| Is This Answer Correct ? | 2 Yes | 14 No |
why cone formation during dissolution?
What impact analysis required for out of calibration and how to proceed
In partical size by malven,which parameter is generlly change for development of dry and wet method.
What is the major difference b/w Drift and Noise
4 Answers Aurobindo, Dr Reddys, Reddy Labs,
what r the parts calibrated in gc
can we use the same detector in HPLC as well GC and what could be the differences we can find in the final chromato graph in any aspects?
what is the difference between IR and FTIR instrument analysis
HOW TO CALIBRATE HPLC OVEN
How to calculate lamp energy in hplc?
what is protocol ? plz Explain me
how to prepare 5 normality of h2so4 solution
why we are calibrating autopolarimeter with quartz control plates