Explain about Low pressure gradient and High Pressure
gradient systems in HPLC?

Answers were Sorted based on User's Feedback



Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / jayesh patel

In low pressure gradient elution mixing of solvent done
before the pumps and in high pressure gradient elution
mixing of solvents done after passing from pumps.

Is This Answer Correct ?    38 Yes 2 No

Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / raghava.mandalapu@gmail.com

Low pressure gradient means Both the solents mix before pump
High pressure gradient means Both the solents mix after pumps

Is This Answer Correct ?    27 Yes 3 No

Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / venkata narayana

Answer:
In a high pressure gradient pump, each different mobile phase is delivered by an individual pump head and then the mobile phases are mixed at the pump outlet. In a low pressure gradient pump, different mobile phases are mixed using a valve before entering the pump head.


As a result of the fact that the low-pressure gradient design uses only one pump head, it is of lower cost. It can also use more types of mobile phase without significant increase of cost.


Since solvent mixing point is much closer to the column head in the high-pressure gradient design, it provides a much faster gradient. This is measured using delay volume. The value can be 50-300 uL for high pressure gradient pump and can be 2 to 3 times larger for a low pressure gradient pump. A small delay volume is important when the analysis time is short or the flow rate is low. If the delay volume is too large, it become impossible to obtain reproducible gradient run since the planed composition cannot reach the column head before a run is finished.

Is This Answer Correct ?    20 Yes 1 No

Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / tariq

HPLC gradient elution systems can be categorized according to their configuration as high-pressure gradient elution systems or low-pressure gradient elution systems.

IN A LOW-PRESSURE GRADIENT ELUTION SYSTEM

(1) Driving the Solenoid (Mobile Phase Mixing Valve)
In a low-pressure gradient elution system, a solenoid valve switches in synchronization with the pump intake stroke to mix multiple mobile phases. The small-capacity plunger pump requires the solenoid valve to switch very frequently and is disadvantageous to making accurate concentration settings.

(2) Effects Due to Mobile Phase Compressibility
No mobile phase intake occurs in the hatched area as the pressure increases, this area gets larger and the concentration error increases.

(3) Correcting Errors Due to Solenoid Valve Response Delay
A time delay occurs from the time of signal input until the solenoid valve operation is complete.

(4) Mechanical Errors
Mechanical errors in the plunger drive mechanism (cams, etc.) affect the concentration accuracy

(5) Accuracy
The main difference between the systems is the ability of the high pressure one to generate more accurate and reproducible gradients at flow rates and within narrow ranges of mobile phase variations.
(6) Tests
In a low-pressure gradient elution system you can’t perform the test of high resolutions. It is used to perform simple protein test with low resolutions
(7) Technology
It is an old technology formerly used
(8) Outgassing
Outgassing of dissolved gases can cause more problems in low-pressure mixing so degassing of mobile phase is essential

(9) Maintenance
The low pressure gradient system required more maintenance.

(10) Ternary Or Quaternary Gradients
The low pressure system will let you run ternary gradients or quaternary if you wish, and let you quickly select up to 4 different solvents. However, you a binary high pressure system with a switch valve that will let you use or select up to four solvents.
(11) Dwell Volume
Dwell volume is the physical volume of liquid between the point of mixing and the column. On a high pressure mixing system, the mixing occurs after the pumps, and usually there is only a small dwell volume. On a low pressure mixing system, the mixing occurs before the pump, and the dwell volume includes the volume in the mixing chamber and the low pressure tubing leading to the pump, in addition to the tubing after the pump. This volume is usually much larger than with high pressure mixing, and can be 1 mL or more.
(12) Gradient Range
The optimum gradient range for the low pressure pump is between 5% to 95% organic. The binary pump shows optimum performance from 0% to 100% organic, which is advantageous for compounds eluting at very low or at very high organic percentage.
(13) Mixing Efficiency
Another advantage for the binary pump is the high mixing efficiency, which is necessary when TFA is used as a modifier.
In the end
HP mixing systems will always outperform LP mixing system under all conditions. Hence two Pumps are most suitable for high performance applications of a gradient system requiring high pressures with pulseless and accurate flow.

Is This Answer Correct ?    3 Yes 1 No

Post New Answer

More Analytical Chemistry Interview Questions

how much wave length of Soduim vapour lamp used in polarimeter

2 Answers   Reddy Labs, Torrent Pharma,


What is vial equlibriation in Gc?

1 Answers  


How much is the standard area for glc analysis

0 Answers  


Which Zone of our country like conducting of Stability studies(Accelerated & Long term stability studies)as per ICH guidelines.

4 Answers  


How would you decide dissolution medium for NCE compound of class I drug

0 Answers  






In dissolution test s1 stage we are getting one bowl 40 % out of 6 bowls what is the root cause you find and how proceed next step? ( Q is 80%)

3 Answers   Reddy Labs,


Please tell me about the pH of Polycaboxylic ether is it in the 5-6 range ever or more than 6

0 Answers   Apex Encon Projects Private Limited, Cipla,


If we have 5 strength which is not dose proportinate and excipients also diffrent in each strength then how we can proceed for Force degradation? and excipient are same but not dose propotinate the how FD?

0 Answers  


How we will diapose UV (detector) lamp?

1 Answers   Bayer,


why use k2cr2o7 , kcl h2so4 in uv calibration

1 Answers   DRL, GVK,


we can use expired sample for validation and analyst qualification?

0 Answers  


how do start hplc method development? how much sample required for developing the method development?

3 Answers   BMS, Cipla, Indoco Remedies, Sun Pharma,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)