Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


How you will select the runtime in HPLC for a drug in
method development.

Answers were Sorted based on User's Feedback



How you will select the runtime in HPLC for a drug in method development...

Answer / ankur dwivedi

There is no any thumb rule that the runtime should be 3
times than main peak, but also depends on the late eluting
impurities in compound; which should be confirmed at the
time of method development. Considering all these, the run
time should be selected.

Ankur Dwivedi
ZCL Chemicals, Ankleshwar

Is This Answer Correct ?    24 Yes 1 No

How you will select the runtime in HPLC for a drug in method development...

Answer / sarma dvv

In the method development, we have to extend the run time 7-
8 times more with analyte retion time. we have to check any
other peaks are eluting. In that run nopeaks are eluted
means accordingly we can fix the run time.
Generally for isocretic 3times more with resect to analyte
peak.

For Gradient we have to extend the organic concentration
maximum possible.

Is This Answer Correct ?    15 Yes 2 No

How you will select the runtime in HPLC for a drug in method development...

Answer / sundar

the runtime will be three times of main analye retention
time.

Is This Answer Correct ?    13 Yes 6 No

How you will select the runtime in HPLC for a drug in method development...

Answer / rahul

There are two three concerns to keep in mind when setting a
run time during method development.
1)Refer the degradation study data of the product.due to
this you come to know about the product beheviour with
method during stress conditions.
2)Confirm the impurity and main compund mass balance test
because of this you will assure about no further elution.
3)Run the HPLC for 5X to the column volume due to this you
can assure about and late elutions.

If with these experiments there is no elution after certain
time you can set the run time after 5 to 10mins after last
peak elution.

Is This Answer Correct ?    4 Yes 0 No

How you will select the runtime in HPLC for a drug in method development...

Answer / ajay vishwakarma

depends on the last possible impurities in compound; which
should be confirmed at the time of method development.and
degradation Study,further Run time 5.0times more W.r.t to
Analyte Peak

Is This Answer Correct ?    4 Yes 2 No

How you will select the runtime in HPLC for a drug in method development...

Answer / uvreddy

depends upon main analyte retention time .

Is This Answer Correct ?    6 Yes 7 No

Post New Answer

More Analytical Chemistry Interview Questions

which effect is mainly responsible for chromatographic separation when silica is the statinary phase at TLC? thanks in advance

1 Answers  


In specificity what parametrs are consider why that parametrs

2 Answers  


When ever we perform system suitability,inject the corresponding standard 6 times why? is thair any reson

4 Answers  


How to interpret XRD graphs against standard and limit for same?

3 Answers   NCL,


why chiral compounds will rotate the plane polarized light in SoR.

4 Answers   GVK,


in which situation require to take incident in validation?

0 Answers  


how to prepare working standard in our laboratory aganist reference standard ?

4 Answers   Merit Biolabs, Unichem,


difference between selectivity and specificity

2 Answers   Troikaa Pharmaceuticals,


What is FFA, PV, and PAV value in oil.

1 Answers   Nestle,


How do you verify the primary,secondary and tertiary amines by IR-Spectroscopy.

2 Answers   GVK,


I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?

0 Answers  


Who discovered X-ray

2 Answers   Huntsman Corporation,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)