Answer / abhijeet gavali

You have two different methods for it
1.0 Use Propyl parabean in methanol and pure methanol
2.0 use 0.05% solution of acetone in water and water

Is This Answer Correct ?    14 Yes 3 No

What is calibration procedure of gredient HPLC pump..

Answer / drena

Mobile phase A is 5.6mg/l of propylparaben in methanol (port
A and B)
Mobile phase B is 100% methanol (port C and D)

Run the test for 20 min by using curve 11 method
sample is Methanol

Acceptance criteria:
A/B/C is between 9.5 % to 10.5
A/B/D is between 9.5 % to 10.5

Is This Answer Correct ?    13 Yes 5 No

What is calibration procedure of gredient HPLC pump..

Answer / kuchela rao athota

in waters gradient program

TIME PUMPA% pUMPB% PUMPC% PUMPD% CURVE
0.00 50% 50% 0 0
2.00 0% 0% 50 50 11
6.00 50% 50% 0% 0 11
10.00 45% 45% 10 0 11
12.00 50% 50% 0 0% 11
14.00 45% 45% 0 10 11
16.00 50% 50% 0 0 11
18.00 50% 50% 0 0 11

Is This Answer Correct ?    8 Yes 6 No

What is calibration procedure of gredient HPLC pump..

Answer / prerna

Gradient Programme for port A and C :
00 : 100
00 : 100
25 : 75
25 : 75
50 : 50
50 : 50
75 : 25
75 : 25
100 : 00
100 : 00

Above programme also run for port B and D

An increasing stepwise piramid will appear. Integrate that
peak and also split each step of the peak.

Find out its height and area of height

By using the formula ( which I don't remeber ), mean flow
of solvent at each step.
Flow : Limit
25 (24-26)
50 (49-51)
75 (74-76)
100 (100)

If the observed value is within the acceptance limit then
that pump passes the test.

Is This Answer Correct ?    10 Yes 10 No

Post New Answer

More Analytical Chemistry Interview Questions

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate.

3 Answers  


please explian me about area normalisation method,diluted standard method,impurity vs impurity and which one has to be selected in method development

0 Answers   Pfizer,


what is the limit of Correlation for IR test compare to the standard and test

3 Answers   Nakoda Chemicals,


Through assay by hplc(anhydrous basis) product got 100.8% .How much the potency of that product.Moisture content is 1.7%,Sulfated ash is 0.02%,Total impurities are 0.3%

2 Answers  


what is method development parameter in HPLC method development

6 Answers   GVK,






what is difirence b/w ROI&ROE

1 Answers   Pfizer,


difference between HPLC & LC-MS

9 Answers   Lambda,


wat is reverse phase in hplc

2 Answers  


What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.

11 Answers   Laurus, Teva API,


What is the diffrence between the performing of LOD ordinary drying at voven and vaccum method.

4 Answers   Apex,


why we use Dichlorobenzene.nitrobezine.t-butyldi sulhate.for calibration of gc Head space

0 Answers   Raks Pharma,


How you collect dissolution media while in stirring condition of a basket or a paddle?

2 Answers   Mylan,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)