Answer / santosh kumar bariki
1. find out wave length where the response of known impurity and api are similar or nearer.
2.find out the concentration where the response of impurity and api can be easily quantitated with accuracy and precise (i.e loq at that wavelength).
3.calculate test concentration such that this loq concentration should be less than 0.05% (reporting threshold).
for example if determined concentration is 1 ppm where it can be easily quantitated with precision and accurate.
than 1 is 0.05% than what should be for 100%. that is 2000 ppm is test concentration.
if all impurity has different response consider impurity which have low response
for more clarification contact 07709180670
| Is This Answer Correct ? | 36 Yes | 1 No |
Answer / sambhaji pawar
Depends on the linearity of the principal peak ,LOD and LOQ of all impurities,peak purity these parameters are important to fix the concentration in rel.sub.development. sambhaji pawar 8169166702
| Is This Answer Correct ? | 0 Yes | 2 No |
Answer / dr.v.gangadhar 07702224651
find out the concentration based on signal to noise ratio of the impurity which is eluted with low responce. S/N should be greater than 30 for Specification level. (0.15% or 0.10%)(Minimum three times higher than the LOQ value and 10 times higher than LOD.
| Is This Answer Correct ? | 2 Yes | 7 No |
why the same column give different retention time in hplc?
Why do we use KCl ,K2Cr207,Toulene & Halonium oxide for the caliberation of UV
What is the reason for using the 900ml of dissolution medium
gc Calibration % RSD failed what is the possible reasons?
acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate.
how to calculate elemental impurities?
how to regulate noise in HPLC?
2 Answers Alkem Labs, Vimta Labs,
For Recent draft guidelines for Dissolution test apparatus calibration, why removing the salicylate tables?
why using buffers in hplc inteed of water
calibrtion procedue of LC-MS
why SGF media used for dissolution . What is the use of it .can we use other media of lower pH instead ot this?
How do you give training to a new chemist