gc Calibration % RSD failed what is the possible reasons?
Answers were Sorted based on User's Feedback
Answer / hiren
First is not column proper installations
Secound is not proper work syrring
Thred is injector septa so much punching
Is This Answer Correct ? | 21 Yes | 0 No |
Answer / yogesh patel
flow rate of gas not proper its cose fluctuation in assay
Is This Answer Correct ? | 5 Yes | 0 No |
Answer / alasyam ramesh
1) Improper stabilisation of column and system.
2) injector septa bleeding
3) partial blockage of detector jet.
4) syring struck due to solvents nature( solvent viscosity).
5) plunger moving and taking improper injection volume.
6) some time may effect room temperature also
Is This Answer Correct ? | 5 Yes | 0 No |
If %RSD is failing in GC calibration means we have to check & replace following things,
1.Injector septa
2.glass wools in liner
3.syringe
Thanks,
B.Gangatharan.
Is This Answer Correct ? | 4 Yes | 0 No |
How precise should be the linearity in HPLC method validation?
Describe your experience with quality assurance and quality control analyses?
11 Answers Aurobindo, Dr Reddys, MSC, Reddy Labs,
What is linearity in HPLC and how will you test?
Is it nessesary all multimedia dissolution require descriminatory?
How much carbon loading in nomalphase columns?
Wath is the differenc between two column HPLC wich have the samme support , lenth , pore size, etc.. But different name .for exemple : Spherisorb / lichrosorb
what is the differance between Analytical and semi micro balance?
What is difference between the working standard and reference standard?
what is differance between volatile and non volatile compound expalin by exaple
why require to add enzymes in capsules dissolution only? is it require to tablet also?
how can you determine a compound acidic or basic?
difference between instrumenent and equipment