Why we more noise in PDA detector compared to UV detector?
Answer / anil
There is no Integration function in the conventional UV-Vis spectrophotometer which
accumulates the signal. For example, the total required time will be 1000 sec. for 1000 data
points and it takes 1 sec. to measure one datum. In this case, all 1000 data have the same signal
to noise ratio (S/N).
But in a PDA instrument which has a 1000 photodiode array, 1000 data points can be
measured in 1 sec. and it would take 1/1000 sec. to achieve the same result obtainable in 1 sec.
in a conventional instrument. Therefore, when the same sample is measured for 1000 sec in a
PDA instrument, the signal is accumulated and is 1000 times greater than when measuring for 1
sec. The noise will be 10001/2
. This means that the S/N ratio is improved by 10001/2
.
This resulting benefit of fast data acquisition is termed Felgett’s S/N Advantage or
Multichannel Advantage
Is This Answer Correct ? | 10 Yes | 1 No |
Basic principle of ESI?
how you start the solubility study?
What is Delay & Dead Volume
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
what is the defferance between the potency and working standard
1 Answers Arbro Pharmaceuticals, Caplin Point, ECOlogics, Matrix, Orchid, Venus Remedies,
what is different between volumetric solution and general solution
what is the meaning of linearity,principle of co-relation coefficient?
what is difference between IR & FTIR ? how we calibarte both ? I want to know the procedure.
If inhouse hplc related substance method is completly diffrent from Usp for finished proďuct with diffrent impurities then how require to prove method equivalecy?
what is mean by assay?
How are PDA Detector work in HPLC
why we are using benzene, anyline in acetic anhdride assay titration method?