IN pH CALIBRATION,CALIBRATION IS DONE FOR pH METER OR
ELECTRODE?HOW WILL YOU JUDGE THAT THE METER TO BE
CALIBRATED WITH 4,7,9.2 BUFFERS?EXPLAIN TEMPERATURE EFFECTS
ON pH.
Answer / prabhu
Hi dear the calibration is also one of the system
suitability test,whether the system working satisfactory or
not physically so when you do the calibration for pH meter
with standard pH buffer solution (which is from NIST-it may
accurate solution for actual pH)and once if your instrument
has calibrated,you can cross verify your insrument against
those pH buffer solution (NIST) and the acceptable range is
+/- 0.02 then you can conclude for ?
once if temperature raises the pH will decrease and its not
all cases
Is This Answer Correct ? | 18 Yes | 5 No |
how to get accurate result for Residue on evaporation in purified water
what is the importence and role of LCMS and LCMS/MS and their applications? which type of compounds are analysis by lcms?
what is dissolution What is unit of KF? what is the significant of pH in HPlc? WHY POTASSIUM HYDROGEN PTHALATE IS USED FOR PERCHLORIC ACID? What is difference btw UV detector & RID detector?
why 0.005 M sulphuric acid is used to calbrate spectrophotometer and why caffiene is used for calibrating hplc
2 Answers GVK, Torrent Pharma,
which are the guidelines for force degradation studies?
what are the important specifications for a Column ?
why are not used methonol in kf titrator to find out water content in ketones?
How to set analytical specification for combination products?
a bottle of concerntrated HCL (36.5) was labelled as 37% (w/w) and density 1.19m/l.What is hcl molarity?
what is the definition of Precision.
What is meaning of audit trail in hplc
3 Answers DRL, Lee Pharma, Raks Pharma,
How can I develope method of dissolution by HPLC OR UV