This is only a briefe answer.
OOT means out of trend.
Definition:
In case of routene analysis any result of quantitative
chemical test that do not fall with in the defined band in
comparison to a specification.
Investigation:
Investigation procedure is same as that of OOS(Out of
specification).
OOt result in qc laboratory
.
Report with in 24 hours to qa&supervisor.
.
phase I laboratory investigation is carried out.
.
If any laboratory error is identified then it is termed as
assiagnable cause(ie system problem identified,sample
preparation,dilution in correct,improper transfer of
solvent,analyst etc etc...).There are many check points are
there to investigate.The investigator will go through the
check points.
.
After assiagnable cause identification corrective and
preventive action will be taken,and order for re testing
the original sample or another on same or another equipment
(if system performance is suspected).The person other than
who obtain the initial results should re test the
sample.After re testing results are to be evaluated and
conclude the investigation.
Re testing plan:2 sample preparetions to be done and analyse
Different different OOT limits are there for different
different tests.
For more details mail me.
| Is This Answer Correct ? | 16 Yes | 0 No |
Answer / manojmalik
Investigation of OOT ,like same as per the OOS
investigation ,Obseved the OOt as per the regrassion
cofficient .based ob 1,2and 3 sigms value .
| Is This Answer Correct ? | 3 Yes | 1 No |
Answer / vamsidhar reddy. b
1. Tabulate the data, calculate the average and standard deviation.
• Use excel spread sheet for average & Standard deviation calculation.
• Standard deviation will be multiplied by 3 to get the 3 sigma (3 )) value.
• Calculate ± 3 Sigma
• Maximum & Minimum limits will be fixed by the formula Average ±3 Sigma
• If the obtained maximum value is above the spec. Limit, specification limit shall be taken into consideration.
• The obtained minimum value may be below the lower spec. limit.
• The maximum and minimum limits as explained above shall be taken as the trend range for upper and lower limits.
• Any obtained value is found to be out of this range will be considered as Out of Trend (OOT) value.
2. Based on preliminary investigation OOT Results should be reconfirmed with same analyst in duplicate.
3. If the result of repeated analysis is in line with the result of 1st analysis, then results of 1st analysis shall be reported and 2nd analysis is considered ‘For confirmation only’ and further investigation is carried out for the cause of OOT result.
4. If the result of repeated analysis is line with previous trend, then result of 2nd analysis shall be reported and no further investigation is carried out with respect to the product.
| Is This Answer Correct ? | 1 Yes | 0 No |
Why we use Caffeine in hplc calibration
What is Impuriteis?
Why the ph limit (1_14) Is used?
why we are doing seal wash and needle wash?
In HPLC when we use Isocratic method, Gradient method in hplc?
what is diffrence between specificity and selecivity?
how much mass should be there in volumetric flask while in preparation of sample for assay?
What is mean by PDR?
how to selecet an exact coloumn for an new molecule development by hplc how to select exact salt as buffer for new molecule development by hplc what is the the process to select the mode of saparation of compoundes by hplc what is the use of ph of buffer what is use of buffer,ph,organic phase,ans methods how the molecules get saparated in coloumn,
why disodium tartarate used in karl-fischer instrument ?.
What is the diffrence between the performing of LOD ordinary drying at voven and vaccum method.
how do you perform reverse enginearing study for drug containing sodium chloride sodium citrate sodium bicarbonate