In the test for control of absorbance in UV calibration why
do we use only potassium dichromate and what is purpose of
taking a specified amount of 57.0-63.0mg?
Answers were Sorted based on User's Feedback
Answer / bhagya lakshmi.k
It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm
| Is This Answer Correct ? | 53 Yes | 4 No |
Answer / sunil.m.kulkarni ( avon/arch p
Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concentration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectrophotometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs
| Is This Answer Correct ? | 36 Yes | 5 No |
Answer / sagadevan
becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,
| Is This Answer Correct ? | 18 Yes | 4 No |
Answer / devidas
it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity
| Is This Answer Correct ? | 18 Yes | 7 No |
Answer / vivek
we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...
| Is This Answer Correct ? | 12 Yes | 5 No |
Answer / srinathsagar
UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)
| Is This Answer Correct ? | 4 Yes | 1 No |
because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm
| Is This Answer Correct ? | 6 Yes | 4 No |
1 % solution is equal to how much ppm
sop of a uv visible spectrophotometer double beam elico model
how do you perform reverse enginearing study for drug containing sodium chloride sodium citrate sodium bicarbonate
What is difference btw UV detector & RID detector
In partical size by malven,which parameter is generlly change for development of dry and wet method.
How to know HPLC column performance?
How they found 1mL of K.F reagent is equivalent to 5mg of water and if we change the composition of K.F reagent, is it can neutralize more amount of water?
Why forced degradation studies are performed in hplc method validation?
why hexadecane is used in calibration of FID IN GC THERE IS NO REMAINING SOLVENTS
What is FFA, PV, and PAV value in oil.
Whya the need of Specificity to do in Analytical Method Validation ?
if peak get problamatic then what require to do?