control of resolution in calibration of IR
Answers were Sorted based on User's Feedback
Answer / nirav intas pharma
control of resolution in IR:
Take spectrum of polystyrine film,
the abs minima at 2870 cm- & abs maxima at 2849.5cm- is
greater than 0.33
the abs minima at 1589 cm- & abs maxima at 1583 cm- is
greater than 0.08
Is This Answer Correct ? | 12 Yes | 0 No |
Answer / kuchela rao athota
MAXIMUM ABSORPTION AT ABPUT 1583CM-1 SHOULD BE GREATER THAN
12%
MAXIMUM ABSORPTION AT ABPUT 2850CM-1 SHOULD BE GREATER THAN
18%
Is This Answer Correct ? | 5 Yes | 0 No |
Answer / arpit vashi
the difference between two percentage transmittance
(2870 cm-1 and 2849.5 cm-1) must be grater than 18%
the difference between two percentage transmittance
(1589 cm-1 and 1583 cm-1) must be grater than 10%
Is This Answer Correct ? | 1 Yes | 0 No |
what r the reasons to get peak telling?
what type of question will ask in the interview of lab chemist.
0 Answers Godavari Biorefineries, Monginis,
How do u differentiate primary
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
what is mean by ambient temperature?
Why we have to take 42gm of sodium hydroxide instead of 40gm for preparation of 1M Sodium hydroxide as per Pharmacopia? Whether it means the formula for calculating molarity was wrong?
5 Answers Claris Otsuka, Ranbaxy,
Why potassium hydrogen pthalate is used for perchloric acid And naoh. Why it is used for both acid and base
6 Answers Vijayasri Chemicals,
Which water can use Hplc analysis
9 Answers BioGenomics, Dr Reddys, Reddy Labs,
How to decided test concentration in GC analysis.
what is specificity?
4 Answers Amneal Pharmaceuticals, Aurobindo,
What is use of LOD and LOQ in validtion and give range and in calculation why we use 3.3* LOD here why we use 3.3
2 Answers Dr Reddys, IPCA, Reddy Labs,
EXPLAIN THE CALCULATION PART OF AN KF TITRATOR CALIBRATION WITH SUITABLE EXAMPLE