Answer / alpesh patel
It means yr column is contaminated & the column needs
proper washing with suitable solvents. You can also
regenarate the column by proper washing with
polar solvent -> semi polar solvent -> non polar solvent ->
semi polar solvent -> polar solvent.
Is This Answer Correct ? | 18 Yes | 4 No |
Answer / mr. bujji reddy kanchi
The peak split involves both mobile phase and stationery
phase. The pH of the mobile phase accuracy depends.
Mainly Stationery phase deterioration. Column has to be
regenerated according to the Column certificate
recommondation solvents.
Is This Answer Correct ? | 6 Yes | 0 No |
Answer / satyasri
if u observe peak splitting,first u go to change(reduce) injection volume(means volume over load)...dn u go for buffer srength(means conc)...dn u go for pH of the buffer...dn only u go for checking of column performance(means whether u hv to check column was washed properly from previos one)........k
Is This Answer Correct ? | 2 Yes | 1 No |
If the sample diluent is not compatable with mobile phase,
it could be the reason for peak splitting. Using mobile
phase as sample diluent would resolve the issue.
Is This Answer Correct ? | 1 Yes | 0 No |
Main cause : That means u r column is contaminated
absorbant ion of impurites in the column.
Solution :
1)Wash calumn:-By polar solvent -> semi polar solvent -> non
polar solvent ->semi polar solvent -> polar solvent.
2)Check column performance
3)Replace column
Is This Answer Correct ? | 0 Yes | 3 No |
Answer / umesh patel
BECOZ OF COLUMN OVERLOADING
DECREASE THE INJECTION VOLUME AND CHECK OUT
ALSO
IT HAPPENS THAT IMPURITIES FROM PREVIOUS INJECTIONS ELUTED
RUN BLANK WITH PROPER SOLVENT AND THEN WITH MOBILE PHASE
Is This Answer Correct ? | 1 Yes | 5 No |
Answer / n.s.r
In Method development time or already validated method first we conform that already validated method it's ok but in development time we go on peak purity and finalise column problem or if any impurities is present in main peak.Thank you ANY WAY YOU WILL answer that question.
Is This Answer Correct ? | 0 Yes | 6 No |
if tech transfer fails whay require to do?
Why sodium lamp is used as source of Polarimeter?
12 Answers College School Exams Tests, Hetero, NCL, Zydus Cadila,
what is the general mechanism of high performance liqiud chromatography?
7 Answers Ajanta Pharma, Amneal Pharmaceuticals,
i set electrode to 7 buffer than i used it to check 4 buffer and 10 buffer if the results are not as expected than is it better to set the buffer again to 4 buffer or is there any other method available please let me know ...
Why are EDTA used to etimation of Ca,Mg etc, in Water analysis?
which lights using during force degradation?
what is the purge flow & how to calculate
what is lod and loq
What is difference between oos
what is mean by nitrosamine impurities?
What is difference between silica used in TLC and HPLC column
4 Answers Cipla, Meghmani Organics,
why we should take dst factor for below 1%moisture samples
0 Answers Ind Swift Laboratories,