How precise should be the linearity in HPLC method validation?
Answers were Sorted based on User's Feedback
Answer / m
Linearity is showing the precisiness of results over the
specification range of LOQ or Lower concentration to 150%
or 200% of specification concentration and it is proved
that your method is precise over perticular concentration
selection range.
| Is This Answer Correct ? | 6 Yes | 2 No |
Answer / premraj behera
Based on the final concentration of the samples/standards
used for the validation of the method, homogeneity and
stability, the linearity plot (Concentration Vs Area) will
be plotted which covers the final concentration with ± 50%
range. so that analyzed concentration of the
samples/standards will fall within these range which is
acceptable.
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / shadab aga
WHAT IS THE SPECIFIC OPTICAL ROTATION OF SUGAR?
| Is This Answer Correct ? | 1 Yes | 1 No |
Principle of HPLC in one sentance
what is the difference between Tailing Factor & Asymmetry
8 Answers Ashish Life Science, BEC Chemicals, Jubilant,
Which Zone of our country like conducting of Stability studies(Accelerated & Long term stability studies)as per ICH guidelines.
what is partion and column chromatography
What is the difference between low pressure gredient and high pressure gredient in hplc instrument?
As per ICH related substances stability trend limit from initial to shelf life
0 Answers Medley Pharmaceuticals, Micro Labs,
How to calculate the signal to noise ratio.
Why chromic acid wash is not recommended by USP for glassware cleaning?
Describe your experience with quality assurance and quality control analyses?
11 Answers Aurobindo, Dr Reddys, MSC, Reddy Labs,
in sampling of raw material which side is pasted sampled labels
If compound Pka is 5 then which buffer you will select ?
What is the definition of RRF